I am resurrecting a G1364C fraction collector. Initialization failed and gave error: missing or wrong needle. LabAdvisor will not move the arm into a maintenance position and is giving error: Unknown Error. Any ideas?
I am resurrecting a G1364C fraction collector. Initialization failed and gave error: missing or wrong needle. LabAdvisor will not move the arm into a maintenance position and is giving error: Unknown Error. Any ideas?
Hi CarolynGCSL , I haved just faced the same issues with my two G1364C fraction collectors in the lab. Did you perhaps find a solution with your device? If so could you share the solution please?
abasdeo everything with my devices installed, I only had not used them for a while. Could you propose a solution?
Here is the older user manual: G1364-90012.book (agilent.com) and starting on page 127 there are instruction on how to do the alignment via the old software or in Lab Advisor. If nothing is happening, try a forced cold start (+) Forced Cold Start Procedure - Articles - LC Portal - Agilent Community
Here is the older user manual: G1364-90012.book (agilent.com) and starting on page 127 there are instruction on how to do the alignment via the old software or in Lab Advisor. If nothing is happening, try a forced cold start (+) Forced Cold Start Procedure - Articles - LC Portal - Agilent Community