Hi
We have Thermo CAD with 1260 Agilent. You need to order analog out kit (it is optional so you might not have it) from Thermo for CAD. Then it is easy to configure with UIB. You will see signal in UIB (don't forget about correctdata rate in UIB).
There is Thermo driver for OL so it is possible now to have CAD controlled by OL. We're in the middle of configuration but it looks promising.