Hey, see page 5 of the UIB manual Agilent 1200 Infinity Series Universal Interface Box User Manual. You will need from the CAD an analog in, then from the UIB II a CAN connection to the LC stack. Then you can reconfigure your LC instrument with the UIB II. To do this, you can use the control panel and go to instruments, then highlight the instrument you want to reconfigure (or create a new instrument), then at the top close connecton and then configure the instrument. In the configure instrument, use the auto configure and type in the IP address or the mac address you are using to communicate with you LC stack. This should then automatically detect the UIB.
Hey, see page 5 of the UIB manual Agilent 1200 Infinity Series Universal Interface Box User Manual. You will need from the CAD an analog in, then from the UIB II a CAN connection to the LC stack. Then you can reconfigure your LC instrument with the UIB II. To do this, you can use the control panel and go to instruments, then highlight the instrument you want to reconfigure (or create a new instrument), then at the top close connecton and then configure the instrument. In the configure instrument, use the auto configure and type in the IP address or the mac address you are using to communicate with you LC stack. This should then automatically detect the UIB.