Hello
Please forgive me as I am still new to this but I have not found the answer to my question.
After injection, the baseline drifts downward but it still shows the peaks. The mobile phase is fresh, the column is less than 6 months old. I just replaced the Deuterium lamp. Is there somewhere else I should start looking to hopefully resolve this issue? Please let me know if you need any more details to possibly help me out.
Thank you very much for any help, I couldn't do what I do without you all
KH