I really need to collect a chromatogram at 360 nm on my DAD. How do I do that?
Find maintenance, best practice, and troubleshooting articles.
I really need to collect a chromatogram at 360 nm on my DAD. How do I do that?
What´s the issue? Just set the wavelength on your CDS and a suitable reference, if needed. If you share more details about the model#, I can point you to the user manual.
Here´s an old, but useful document about DAD optimization:
What´s the issue? Just set the wavelength on your CDS and a suitable reference, if needed. If you share more details about the model#, I can point you to the user manual.
Here´s an old, but useful document about DAD optimization:
Honestly, I doubt it is a G1316A - that would be a column thermostat, which can only heat up or cool down columns.
For the reference wavelength, there´s a guideline on page 12 of the DAD optimization document. There is no need to use a reference, but often it´s beneficial. Make sure the reference WL is well separated from the signal WL, as shown in the document. There are many pages on the topic "reference".
Thanks so much. It is a G1316A 1260 TCC. So I have to use a different
reference wavelength? Should it be lower or higher than 360 nm?