How to upgrade MS source to Extractor 350?

Hi,

I have a 5977 MS with SST source. I'm thinking of upgrading to Xtr 350. The MS can support it (There's an extra jack on the board to connect the extractor lens), but I was wondering if I need to make changes to the software or run the Parts Finder after installing the source. Any help would be greatly appreciated!

Thank you!

Ehsan

Parents
  •   What is the model number listed along with the serial number label on your system?  For example, this one is a G7077C, a 5977C system with a turbomolecular high vaccum pump and an extractor ion source.

  • Hi  ,

    Thank you for your attention! The model numberd is G7080B with a diffusion oil pump.

  • The lower the injection port pressure, the less realistic is the vapor volume calculator. It's not perfection in there and 82% is too much.  With Chloroform you'd want the initial oven temperature to be 40° C or even lower. You say your compounds are very light, so reducing it to even 35° C might help chromatographic peak shapes.

    Pulsed splitless.30×250×.25 at 1 ml/min is your normal column flow with the inlet at 7.2892 psi at 45°C.  The diffusion pump in your system has a maximum allowable flow of 2 ml/min and the gas calculator says 16.497 psi to get that, so a pulse of 16 psi until 0.5 minutes and then purge flow 50 to split vent at 1.00 minutes would be a good starting point.  Then have gas saver on 25 ml/min at 2.00 minutes. 2 µL of chloroform at 16 psi and 260° C inlet gives 52% expansion, much better, much more realistic/safe/usable.

    The split splitless and multimode inlets require a minimum of at least 15 ml/min of total flow to function properly.  That 15 ml/min is not really enough to eliminate the possibility of air diffusion into the system, so I recommend 25 ml/min at all times.    If you need to save helium, Agilent sells a Helium Conservation Module that will let you switch to nitrogen whenever the system is not being used.

    You can scan from 10 m/z if you'd like as long as you understand that the TIC baseline will include 18, 28, and 32. 

    ANY 32 - oxygen -- in your system degrades the column and oxidizes the ion source requiring more cleaning. 

  • Thank you for your time and valuable information Paul! I really appreciate it! I am aware of water (18), N2 (28), and O2(32) and constantly check their presence.

    Best!

    Ehsan

  • Hi  

    Just wanted to ask a question out of curiosity. We are located in Fort Collins, 5,000 ft above sea level. So, the atmospheric pressure would be ~12.2 psi here. Is it possible to put inlet pressure at 12.5 psi along with a low septum purge (e.g. 3 ml/min) to prevent air from diffusing to GC when GC is in standby instead of having septum purge of 50ml/min? In this case column flow would be 1.48 ml/min for my setup.

    Thanks!

    Ehsan

  • For standby method I'd recommend inlet pressure ~5 psi.  This is gauge pressure so it'll work to prevent air diffusing into the inlet if there are any small leaks.  Keep in splitless mode, set purge flow to 20 mL/min, septum purge to 3 mL/min.

  • Keep constant flow 1.2 ml/min into the MS and do a Pulsed Splitless injection, tips in the help section of the Operating manual.  Your system's diffusion high vacuum pump can only handle 2.0 ml/min maximum so do not exceed that ever.

    If you have a 30m × 0.25mmid column installed, try these parameters.

    Oven initial temp 45° C for 1 minute at least.

    Pulsed splitless, 18.6 psi pulse for 0.5 minutes, then purge flow to split vent 50 ml/min at 1 minute. Gas Saver ON, 20 ml/min at 2 minutes.

    What is your injection solvent?  You can use the vapor volume calculator in the acquisition software to see if more sample will fit into the liner.

Reply
  • Keep constant flow 1.2 ml/min into the MS and do a Pulsed Splitless injection, tips in the help section of the Operating manual.  Your system's diffusion high vacuum pump can only handle 2.0 ml/min maximum so do not exceed that ever.

    If you have a 30m × 0.25mmid column installed, try these parameters.

    Oven initial temp 45° C for 1 minute at least.

    Pulsed splitless, 18.6 psi pulse for 0.5 minutes, then purge flow to split vent 50 ml/min at 1 minute. Gas Saver ON, 20 ml/min at 2 minutes.

    What is your injection solvent?  You can use the vapor volume calculator in the acquisition software to see if more sample will fit into the liner.