Peak saturation with calibration curve

Hello,

I am trying to increase the range of my calibration curve but I have problems with saturation. First, if the integration list says the peak is saturated, does that mean the peak area is unreliable to use in the calibration curve? In the attached image, the red peak is the standard extract with 10:1 split ratio, the blue peak is the same extract with 50:1 split ratio, and the black peak is 5x diluted extract with 50:1 split ratio. According to the integration list only the black peak is saturated even though we know less analyte has been injected. How does this saturation relate to concentration? I am also assuming the saturation has to do with the number of data points at the peak. Are there certain MS parameters I should adjust to eliminate the saturation? I am also interested in advice to reduce the tailing as concentration decreases.

Parents
  • The maximum abundance per ion on Agilent Single Quadrupole GCMS instruments is approximately 8.4 × 106, To stay within the linear range of the electron multiplier it is recommended to stay at or below 2.0 × 106 for SQ.

    This large peak is 54,000,000 counts tall. You need to further dilute your samples, dramatically increase your split ratio, inject less sample, or do all three.  You want your smallest calibration peak to be maybe 10,000 counts or less tall. The smallest peak shown is 2,000,000. The smallest peak of interest only needs to be about three to five times taller than the background noise to be valid.

    The less you inject of the lowest concentration and cleanest samples possible will require less inlet, column, and ion source maintenance.

     

    What is the sample? What are the GC parameters?  Please upload a copy of the acqmeth.txt file that is located underneath your method's ****.M subdirectory.

    Also older versions of MassHunter Acquisition and old MSD Chemstation sometimes had trouble with data files names longer than 40 characters and this one is already 42 characters long.

  • Thanks for the response. 

    The sample I am measuring is methyl methacrylate. In the attached image, the black peak (measured in scan mode) is still saturated even though the counts are relatively lower. The red peak is the EIC for 100.0 ion, which we use for quantification, and it is not saturated. If the peak is saturated in scan, but the EIC for quant ion is not saturated is it okay to use the peak areas for calibration curve? Is any "saturated" peak reliable for calibration?

    We also noticed that smoothing the chromatogram eliminates the saturation issue, but I am not familiar with smoothing parameters. Some of the saturated peaks have flat tops prior to smoothing, which is why I thought increasing the MS sampling may help.

                           INSTRUMENT CONTROL PARAMETERS:    1
                           -----------------------------------
    
          Wed Feb 14 11:37:32 2024
    
    Control Information
    ------- -----------
    
    Sample Inlet             : GC
    Injection Source         : External Device
    Mass Spectrometer        : Enabled
    
    Injection Location:  Front
    
    
     No Sample Prep method has been assigned to this method.
    
    
    GC
    GC Summary
    Run Time                                     22.133 min
    Post Run Time                                0 min
    
    Oven
    Temperature
    Setpoint                                     On
    (Initial)                                    32 °C
    Hold Time                                    7 min
    Post Run                                     70 °C
    Program
    #1 Rate                                      10 °C/min
    #1 Value                                     70 °C
    #1 Hold Time                                 0 min
    #2 Rate                                      30 °C/min
    #2 Value                                     320 °C
    #2 Hold Time                                 3 min
    
    
    Equilibration Time                           0.25 min
    Max Temperature                              320 °C
    Maximum Temperature Override                 Disabled
    Slow Fan                                     Disabled
    
    Front SS Inlet He
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    Mode                                         Split
    Heater                                       On    250 °C
    Pressure                                     On    17.915 psi
    Total Flow                                   On    91.23 mL/min
    Septum Purge Flow                            On    3 mL/min
    Pre-Run Flow Test                            Off
    Gas Saver                                    On    20 After 3 min mL/min
    Split Ratio                                  50 :1
    Split Flow                                   86.5 mL/min
    Liner                                        Agilent 5190-2293: 900 μL (Splitless, single taper, wool, Ultra )
    
    Back PTV Inlet He
    Mode                                         Sleep
    Pressure                                     Off
    Septum Purge Flow                            Off
    Pre-Run Flow Test                            Off
    Liner                                        Agilent G1544-80730: 675 μL (Direct connect, single taper, bottom )
    
    Column
    Column #1
    Flow
    Setpoint                                     On
    (Initial)                                    1.73 mL/min
    Post Run                                     0.57353 mL/min
    
    Column Information                           Agilent
    Restek-Rxi5SilMS                             
    Temperature Range                            -60 °C—350 °C (350 °C)
    Dimensions                                   30 m x 250 μm x 0.25 μm
    Column lock                                  Unlocked
    In                                           Front SS Inlet He
    Out                                          PSD 2 
    (Initial)                                    32 °C
    Pressure                                     17.915 psi
    Flow                                         1.73 mL/min
    Average Velocity                             34.773 cm/sec
    Holdup Time                                  1.4379 min
    Control Mode                                 Constant Flow
    
    Column #2
    Pressure
    Setpoint                                     On
    (Initial)                                    2 psi
    Post Run                                     10 psi
    
    Temperature Range                            -1 °C—325 °C (325 °C)
    Dimensions                                   5 m x 200 μm x 0 μm
    Column lock                                  Unlocked
    In                                           PSD 2 He
    Out                                          MSD 
    (Initial)                                    32 °C
    Pressure                                     2 psi
    Flow                                         1.5224 mL/min
    Average Velocity                             109.14 cm/sec
    Holdup Time                                  0.076352 min
    Control Mode                                 Constant Pressure
    
    Column Outlet Pressure                       0 psi
    
    Front Detector FID
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    Makeup                                       He
    Heater                                       On    300 °C
    H2 Flow                                      Off
    Air Flow                                     Off
    Makeup Flow                                  Off
    Carrier Gas Flow Correction                  Constant Makeup and Fuel Flow
    Flame                                        Off
    
    Detector Evaluation
    Perform Detector Evaluation Test             Off
    Signal Selected                              Signal #1: None
    Checkout Sample                              None
    
    
    MSD Transfer Line
    Temperature
    Setpoint                                     On
    (Initial)                                    320 °C
    
    
    
    Aux EPC 1,2,3
    Aux EPC 1 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    Aux EPC 2 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    Aux EPC 3 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    
    PSD 2
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    PSD Purge                                    On    0 mL/min
    PSD 2 He                                     Supplies Column 2
    
    Signals
    Signal #1: 
    Description                                  None
    
    Signal #2: 
    Description                                  None
    
    Signal #3: 
    Description                                  None
    
    Signal #4: 
    Description                                  None
    
    Signal #5: 
    Description                                  None
    
    Signal #6: 
    Description                                  None
    
    Signal #7: 
    Description                                  None
    
    Signal #8: 
    Description                                  None
    
    
    
    
    
    
                                GERSTEL MAESTRO
    
    
    SYSTEM SETTINGS
       Maestro Runtime         : 22.13 min
       GC Cool Down Time       : 5.00 min
       Cryo Timeout            : 15.00 min
    
    
    
                                   GERSTEL CIS
    
    
       CIS                     : not used  -  use these parameters if it becomes 'used'
    
    CRYO COOLING
       Cryo Cooling            : used
    
    TEMPERATURE PROGRAM
       Heater Mode             : Standard
    
       Liner Type              : 013882-010-00 glass liner CIS4, baffled, non deactivated
       Liner Max Temperature   : 400 'C
    
       Initial Temperature     : -10 'C
       Equilibration Time      : 0.50 min
       Initial Time            : 0.50 min
       Ramp 1                 
           Rate                : 12.00 'C/s
           End Temp            : 325 'C
           Hold Time           : 5.00 min
       Ramp 2                 
           Rate                : 0.0 'C/s
    
       Low Split               : not used
    
    
    
                                   GERSTEL MPS PREP
    
    
       Sample Prep             : used
    
    PREP ACTION LIST
    
         Action          Method/Value          Source                Destination
         =============   ===================   ===================   ===================
    
    
    
    
    
    
                            GERSTEL MPS Liquid Injection
    
       Syringe                 : 10ul 65mm USM
    
    SAMPLE PARAMETERS
       Sandwich                : not used
    
       Inj. Volume             : 2.0 uL
       Air Volume below        : 0.0 uL
    
       Inj. Speed              : 50.00 uL/s
       Fill Volume             : 5.0 uL
       Fill Strokes            : 5
    
       Fill Speed              : 2.00 uL/s
       Viscositiy Delay        : 0 s
       Eject Speed             : 50.00 uL/s
    
       Pre Inj. Delay          : 0 s
       Post Inj. Delay         : 0 s
    
       Inj. Penetration        : 40.00 mm
       Sample Tray Type        : VT54/2-CVM
       Bottom Sense            : used
       Fast Injection          : used
    
    CLEANING PARAMETERS
       Preclean Sample         : 2
    
       Wash Station 1          : Pos1@Standard Wash 1
       Preclean Solv.1         : 2
       Postclean Solv.1        : 6
       Clean Volume Solv.1     : 90 %
       Fill Speed Solv.1       : 5.00 uL/s
       Viscosity Delay Solv.1  : 0 s
       Eject Speed Solv.1      : 50.00 uL/s
       Information Solv.1      : Hexane
    
       Wash Station 2          : Pos2@Standard Wash 1
       Preclean Solv.2         : 0
       Postclean Solv.2        : 0
       Clean Volume Solv.2     : 70 %
       Fill Speed Solv.2       : 5.00 uL/s
       Viscosity Delay Solv.2  : 0 s
       Eject Speed Solv.2      : 50.00 uL/s
    
    MS Information
    -- -----------
    
    
    General Information
    ------- -----------
    
    
    Acquisition Mode         : Scan
    Solvent Delay (minutes)  : 3.3
    Tune file                : D:\MassHunter\GCMS\1\5977\etune.u
    EM Setting mode Gain     : 5.000000
    
    Normal or Fast Scanning  : Normal Scanning
    Trace Ion Detection      : Off
    Run Time (if MS only)    : 23.67 minutes 
    
    [Scan Parameters]
    Start Time               : 3.3
    Low Mass                 : 40
    High Mass                : 300
    Threshold                : 300
    A/D Samples:             : 4
    
    
    [MSZones]
    
    MS Source                   : 230 C   maximum 250 C
    MS Quad                     : 150 C   maximum 200 C
    
    Timed Events
    ----- ------
    Number Events= 1
    
    
    [Timed MS Detector Table Entries]
    
    Time (min)                 State (MS On/off)
    10.00                  off
    
    
    
    
    
                         END OF MS ACQUISITION PARAMETERS
    
    
                            TUNE PARAMETERS for SN: US2018M021
                            ---------------------------------
    
     Trace Ion Detection is OFF.
    
        34.593   :   EMISSION     
        70.007   :   ENERGY       
         3.790   :   REPELLER     
        89.822   :   IONFOCUS     
        20.181   :   ENTRANCE_LENS
      1669.658   :   EMVOLTS      
                                  2061.7  : Actual EMV 
                                   5.00   : GAIN FACTOR 
      1872.000   :   AMUGAIN      
       133.875   :   AMUOFFSET    
         1.000   :   FILAMENT     
         0.000   :   DCPOLARITY   
         6.990   :   ENTLENSOFFSET
         9.502   :   Ion_Body     
         4.006   :   EXTLENS      
      -250.000   :  MASSGAIN        
       -31.000   :  MASSOFFSET      
    
                            END OF TUNE PARAMETERS
                            ----------------------
    
    
    
                          END OF INSTRUMENT CONTROL PARAMETERS
                          ------------------------------------
    

  • There's a lot going on.  You have Column 1 in Constant Flow to a PSD in Constant Pressure with PSD Purge at 0 ml/min.  Is there a purged splitter plate installed or a Purged Ultimate Union?  The flow into the MS changes throughout the run due to the PSD Constant Pressure setpoint, so at the beginning of the run when the Oven Temp is 32°, the flow into the MS is a bit high at 1.5524 ml/min but at the end of the run the flow is only 0.498 ml/min.  The analytical column constant flow is 1.4379 - so if there is no other place for it to go like to an FID through a splitter, at the end of the run the excess 0.94 ml/min has to go somewhere, so it is going out the PSD purge line and contaminating the PSD flowpath.  This needs some investigation.

    Here are some observations and recommendations.

  • We have the front inlet connected to column> CFT Device> MS transfer line. Our CFT has a purge/vent valve. We have not noticed any carryover or contamination between runs so we are not immediately worried about the purge line, but would you recommend connecting the 2nd column to FID in the future? 


    Referring to the figures-

    The red peak was injected under the previous method. Saturation did not occur. Then we changed to the splitless liner and re-injected (black peak), and the signal increased with saturation. Then we applied some of your recommendations to the method (eqb time, injection temp, and A/D samples) and re-injected (green peak) and again we see saturation. Why is the green peak saturated when it is the lowest abundance? Is there another parameter besides abundance which determines saturation, like the peak width?

    Red peak- injected using the previously described method

    Black peak- Changed from splitless liner to split liner (5190-2295), injected with the same method

    Green peak- 2295 split liner and some recommendations (equilibration time, injection temp, and A/D sampling)

                          INSTRUMENT CONTROL PARAMETERS:    1
                          -----------------------------------
    
          Thu Feb 29 11:53:58 2024
    
    Control Information
    ------- -----------
    
    Sample Inlet             : GC
    Injection Source         : External Device
    Mass Spectrometer        : Enabled
    
    Injection Location:  Front
    
    
     No Sample Prep method has been assigned to this method.
    
    
    GC
    GC Summary
    Run Time                                     18.8 min
    Post Run Time                                0 min
    
    Oven
    Temperature
    Setpoint                                     On
    (Initial)                                    32 °C
    Hold Time                                    7 min
    Post Run                                     70 °C
    Program
    #1 Rate                                      10 °C/min
    #1 Value                                     70 °C
    #1 Hold Time                                 0 min
    #2 Rate                                      30 °C/min
    #2 Value                                     280 °C
    #2 Hold Time                                 1 min
    
    
    Equilibration Time                           2 min
    Max Temperature                              320 °C
    Maximum Temperature Override                 Disabled
    Slow Fan                                     Disabled
    
    Front SS Inlet He
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    Mode                                         Split
    Heater                                       On    250 °C
    Pressure                                     On    17.915 psi
    Total Flow                                   On    91.23 mL/min
    Septum Purge Flow                            On    3 mL/min
    Pre-Run Flow Test                            Off
    Gas Saver                                    On    20 After 3 min mL/min
    Split Ratio                                  50 :1
    Split Flow                                   86.5 mL/min
    Liner                                        Agilent 5190-2293: 900 μL (Splitless, single taper, wool, Ultra )
    
    Back PTV Inlet He
    Mode                                         Sleep
    Pressure                                     Off
    Septum Purge Flow                            Off
    Pre-Run Flow Test                            Off
    Liner                                        Agilent G1544-80730: 675 μL (Direct connect, single taper, bottom )
    
    Column
    Column #1
    Flow
    Setpoint                                     On
    (Initial)                                    1.73 mL/min
    Post Run                                     0.57353 mL/min
    
    Column Information                           Agilent
    Restek-Rxi5SilMS                             
    Temperature Range                            -60 °C—350 °C (350 °C)
    Dimensions                                   30 m x 250 μm x 0.25 μm
    Column lock                                  Unlocked
    In                                           Front SS Inlet He
    Out                                          PSD 2 
    (Initial)                                    32 °C
    Pressure                                     17.915 psi
    Flow                                         1.73 mL/min
    Average Velocity                             34.773 cm/sec
    Holdup Time                                  1.4379 min
    Control Mode                                 Constant Flow
    
    Column #2
    Pressure
    Setpoint                                     On
    (Initial)                                    2 psi
    Post Run                                     10 psi
    
    Temperature Range                            -1 °C—325 °C (325 °C)
    Dimensions                                   5 m x 200 μm x 0 μm
    Column lock                                  Unlocked
    In                                           PSD 2 He
    Out                                          MSD 
    (Initial)                                    32 °C
    Pressure                                     2 psi
    Flow                                         1.5224 mL/min
    Average Velocity                             109.14 cm/sec
    Holdup Time                                  0.076352 min
    Control Mode                                 Constant Pressure
    
    Column Outlet Pressure                       0 psi
    
    Front Detector FID
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    Makeup                                       He
    Heater                                       On    300 °C
    H2 Flow                                      Off
    Air Flow                                     Off
    Makeup Flow                                  Off
    Carrier Gas Flow Correction                  Constant Makeup and Fuel Flow
    Flame                                        Off
    
    Detector Evaluation
    Perform Detector Evaluation Test             Off
    Signal Selected                              No Signal Selected
    Checkout Sample                              None
    
    
    MSD Transfer Line
    Temperature
    Setpoint                                     On
    (Initial)                                    320 °C
    
    
    
    Aux EPC 1,2,3
    Aux EPC 1 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    Aux EPC 2 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    Aux EPC 3 N2
    Pressure
    Setpoint                                     Off
    (Initial)                                    10 psi
    Post Run                                     10 psi
    
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    
    
    PSD 2
    Excluded from Readiness                      ***Excluded from Affecting GC's Readiness State***
    PSD Purge                                    On    0 mL/min
    PSD 2 He                                     Supplies Column 2
    
    Signals
    Signal #1: 
    Description                                  None
    
    Signal #2: 
    Description                                  None
    
    Signal #3: 
    Description                                  None
    
    Signal #4: 
    Description                                  None
    
    Signal #5: 
    Description                                  None
    
    Signal #6: 
    Description                                  None
    
    Signal #7: 
    Description                                  None
    
    Signal #8: 
    Description                                  None
    
    
    
    
    
    
                                GERSTEL MAESTRO
    
    
    SYSTEM SETTINGS
       Maestro Runtime         : 22.13 min
       GC Cool Down Time       : 5.00 min
       Cryo Timeout            : 15.00 min
    
    
    
                                   GERSTEL CIS
    
    
       CIS                     : not used  -  use these parameters if it becomes 'used'
    
    CRYO COOLING
       Cryo Cooling            : used
    
    TEMPERATURE PROGRAM
       Heater Mode             : Standard
    
       Liner Type              : 013882-010-00 glass liner CIS4, baffled, non deactivated
       Liner Max Temperature   : 400 'C
    
       Initial Temperature     : -10 'C
       Equilibration Time      : 0.50 min
       Initial Time            : 0.50 min
       Ramp 1                 
           Rate                : 12.00 'C/s
           End Temp            : 325 'C
           Hold Time           : 5.00 min
       Ramp 2                 
           Rate                : 0.0 'C/s
    
       Low Split               : not used
    
    
    
                                   GERSTEL MPS PREP
    
    
       Sample Prep             : used
    
    PREP ACTION LIST
    
         Action          Method/Value          Source                Destination
         =============   ===================   ===================   ===================
    
    
    
    
    
    
                            GERSTEL MPS Liquid Injection
    
       Syringe                 : 10ul 65mm USM
    
    SAMPLE PARAMETERS
       Sandwich                : not used
    
       Inj. Volume             : 2.0 uL
       Air Volume below        : 0.0 uL
    
       Inj. Speed              : 50.00 uL/s
       Fill Volume             : 5.0 uL
       Fill Strokes            : 5
    
       Fill Speed              : 2.00 uL/s
       Viscositiy Delay        : 0 s
       Eject Speed             : 50.00 uL/s
    
       Pre Inj. Delay          : 0 s
       Post Inj. Delay         : 0 s
    
       Inj. Penetration        : 40.00 mm
       Sample Tray Type        : VT54/2-CVM
       Bottom Sense            : used
       Fast Injection          : used
    
    CLEANING PARAMETERS
       Preclean Sample         : 2
    
       Wash Station 1          : Pos1@Standard Wash 1
       Preclean Solv.1         : 2
       Postclean Solv.1        : 6
       Clean Volume Solv.1     : 90 %
       Fill Speed Solv.1       : 5.00 uL/s
       Viscosity Delay Solv.1  : 0 s
       Eject Speed Solv.1      : 50.00 uL/s
       Information Solv.1      : Hexane
    
       Wash Station 2          : Pos2@Standard Wash 1
       Preclean Solv.2         : 0
       Postclean Solv.2        : 0
       Clean Volume Solv.2     : 70 %
       Fill Speed Solv.2       : 5.00 uL/s
       Viscosity Delay Solv.2  : 0 s
       Eject Speed Solv.2      : 50.00 uL/s
    
    MS Information
    -- -----------
    
    
    General Information
    ------- -----------
    
    
    Acquisition Mode         : Scan
    Solvent Delay (minutes)  : 3.3
    Tune file                : D:\MassHunter\GCMS\1\5977\etune.u
    EM Setting mode Gain     : 5.000000
    
    Normal or Fast Scanning  : Normal Scanning
    Trace Ion Detection      : Off
    Run Time (if MS only)    : 23.67 minutes 
    
    [Scan Parameters]
    Start Time               : 3.3
    Low Mass                 : 40
    High Mass                : 250
    Threshold                : 300
    A/D Samples:             : 2
    
    
    [MSZones]
    
    MS Source                   : 230 C   maximum 250 C
    MS Quad                     : 150 C   maximum 200 C
    
    Timed Events
    ----- ------
    Number Events= 1
    
    
    [Timed MS Detector Table Entries]
    
    Time (min)                 State (MS On/off)
    10.00                  off
    
    
    
    
    
                         END OF MS ACQUISITION PARAMETERS
    
    
                            TUNE PARAMETERS for SN: US2018M021
                            ---------------------------------
    
     Trace Ion Detection is OFF.
    
        34.593   :   EMISSION     
        70.007   :   ENERGY       
         3.989   :   REPELLER     
        89.822   :   IONFOCUS     
        20.181   :   ENTRANCE_LENS
      1668.925   :   EMVOLTS      
                                  2059.7  : Actual EMV 
                                   5.00   : GAIN FACTOR 
      1875.000   :   AMUGAIN      
       133.563   :   AMUOFFSET    
         1.000   :   FILAMENT     
         0.000   :   DCPOLARITY   
         6.990   :   ENTLENSOFFSET
         9.502   :   Ion_Body     
         4.106   :   EXTLENS      
      -251.000   :  MASSGAIN        
       -31.000   :  MASSOFFSET      
    
                            END OF TUNE PARAMETERS
                            ----------------------
    
    
    
                          END OF INSTRUMENT CONTROL PARAMETERS
                          ------------------------------------
    

  • Saturation is reported by ion, not by TIC abundance.

    The maximum possible abundance per ion on SQ instruments is approximately 8.4 × 10^6.
    To stay within the linear range of the electron multiplier it is recommended to be <= 2.0 × 10^6 per ion.

    Please inject less sample, maybe 0.3µL, and/or increase the split ratio, like up to 250:1 or 500:1.

Reply Children
No Data
Was this helpful?