When replacing the PFTBA solution in the calibration vial, the manual says to turn the analyzer off? Does anyone know exactly what this means? Do I need to vent and turn the power off?
When replacing the PFTBA solution in the calibration vial, the manual says to turn the analyzer off? Does anyone know exactly what this means? Do I need to vent and turn the power off?
If the valve is closed why do you need to vent it? Why don't I see PFTBA in every sample then. Instructions don't say to vent it.
Hi all, garyb007 is correct. As long as you are not scanning with the MSD, you can replace PFTBA and then follow the procedure to purge the cal valve.
Hi all, garyb007 is correct. As long as you are not scanning with the MSD, you can replace PFTBA and then follow the procedure to purge the cal valve.