Error in autotune! Adjust Peak Width Failed

Good day Agilent community. We encountered an error in autotuning our GC-QQQ 7890B, in which the following error was stated: "Error in autotune! Adjust Peak Width Failed". My supervisor has tried numerous of other parameters such as doing a manual autotune and we found out the problem arises at MS1, due to check tune indicating that the low, mid and high Isotope mass ratio was a bit outside the parameters. What could possibly be the problem with MS1 and is there any way to fix this before contacting a technician for a full service. Thank you all kindly.

*Edit* The peaks look like baseline/noise and are not resolved during autotuning. My supervisor was able to get our 69 peak resolved during manual tuning and playing around with the quad1 DC voltages however, the peaks went back noise/baseline once we tried autotuning again.

Parents
  • Please share the autotune report. This is the checktune which is not as helpful for troubleshooting.  

  • Hi Paul, thank you for your reply. I can't attach a full autotune report due to it failing mid run. I can however add the text file it generated.

    --------------------------------------------------------------------------------
    Autotune - Starting - 4/19/2023 1:07:20 PM
          Software Revision MassHunter GC/MS Acquisition B.07.05.2479
          PC Hostname DESKTOP-QRSMVU9
          GCQQQ 7010B US1718W001 DSP: 7000.3509, qqqServer: G.7000.058-RUN
          Initial tune file D:\MassHunter\GCMS\1\7010\atunes.eihs.tune.xml
    
    --------------------------------------------------------------------------------
    Tune parameters
      MS1DC 6.4
      MS1PostFilterDC 6.4
      MS1 Mass Axis Offset Dynamic
        enabled True
        69, -1.774, 219, -1.737, 264, -1.762, 414, -1.786, 502, -1.779, 
      MS1MassGain -0.42
      MS1 Width Offset Dynamic
        enabled True
        69, -0.015, 219, -0.002, 264, -0.005, 414, -0.008, 502, -0.013, 
      MS1AMUGain 20.9
      MS1Heater 150
      MS2DC -4.6
      MS2PreFilterDC -14.6
      MS2 Mass Axis Offset Dynamic
        enabled True
        69, -1.882, 219, -1.798, 264, -1.836, 414, -1.851, 502, -1.883, 
      MS2MassGain 10.61
      MS2 Width Offset Dynamic
        enabled True
        69, -0.016, 219, -0.006, 264, -0.005, 414, -0.006, 502, -0.011, 
      MS2AMUGain 13.7
      MS2Heater 150
      CCEntranceDC 7.4
      CCHexapoleDC 6.4
      CCHexapoleRFPeak 400
      CCHexapoleDeltaV -5.0
      CCExitDC 0.4
      CCGasOn 1
      Iris Dynamic
        enabled True
        69, -3.5, 219, -12, 264, -16.5, 414, -26, 502, -30.5, 1050, -58.5, 
      HED -10.0
      EMV 1478
      CollisionEnergy 0
      SourceHeater 230
      Repeller 16.9
      IonBody 19.8
      IonFocus -220.0
      Entrance Lens Dynamic
        enabled True
        69, -15.59995, 264, -16.99995, 414, -19.79996, 502, -21.99997, 
      Filament 2
      Emission 100.0
      ElecEnergy 70
      Extractor Dynamic
        enabled True
        69, 14.60002, 264, 14.90002, 502, 14.60002, 
      SensitivityMode 1
      EinzelLens1 0.0
      EinzelLens2 -12.0
      GasSelect B
      GasFlow 0
      FastScanOffset -4.0
      H2Flow 0.00
    --------------------------------------------------------------------------------
    
    --------------------------------------------------------------------------------
    Acquisition Settings
        MS1ScanMode = TTI
        MS1TTICutoff = 75
        MS1SIMIon = 264
        MS1ScanLowMass = 35
        MS1ScanHighMass = 650
        MS1ScanStep = 0.1
        MS1ProfileMasses = 69.00, 10.00
        MS1ProfileMasses = 264.00, 10.00
        MS1ProfileMasses = 502.00, 10.00
        MS2ScanMode = Profile
        MS2TTICutoff = 50
        MS2SIMIon = 264
        MS2ScanLowMass = 35
        MS2ScanHighMass = 650
        MS2ScanStep = 0.1
        MS2ProfileMasses = 69.00, 10.00
        MS2ProfileMasses = 264.00, 10.00
        MS2ProfileMasses = 502.00, 10.00
        ProfileSamples = 256
        ProfileAverage = 1
        ProfileStep = 0.1
        ScanSamples = 16
        ScanAverage = 3
        ScanThreshold = 0
        ScanIsCentroided = True
        SIMDwellTime = 20
        CaptureSamples = 32000
        CaptureDwellTime = 0.016
    --------------------------------------------------------------------------------
          Doing EI HES autotune
          Repeller set to 12
          IonBody set to 10
          IonFocus set to -180
          Extractor set to 12
          EntLens set to -20
          MS1PostFilterDC set to 5
          Iris set to -35
          Lens 1 Offset set to -10
          Lens 2 set to -25
          Filament Select 2
    
    --------------------------------------------------------------------------------
    Find Peaks, Quad 1 - Starting - 4/19/2023 1:07:20 PM
    --------------------------------------------------------------------------------
          Find Peaks mass 68.08, abundance 121.2, peak width 7.29
    
    --------------------------------------------------------------------------------
    Check Abundance, Quad 1 - Starting - 4/19/2023 1:07:29 PM
          Mass 69.00, Limits 100000 to 2000000
                 EMV,      mass, abundance, peak width
                1478,     66.86,        118,      9.86
                1678,     72.01,        589,      8.65
                1878,     69.13,       2728,      4.88
                2078,     69.19,       9380,      9.62
                2278,     68.98,      25904,      4.94
                2478,     69.12,      71019,     11.93
                2678,     72.31,     180491,      9.30
    Check Abundance, Quad 1 - Finished
          Find Peaks mass 69.18, abundance 194208.2, peak width 8.52
          Find Peaks mass 498.49, abundance 143799.7, peak width 10.79
          Find Peaks mass 70.29, abundance 173434.9, peak width 8.86
          Find Peaks mass 497.63, abundance 171518.2, peak width 50.83
    
    --------------------------------------------------------------------------------
    Check Abundance, Quad 1 - Starting - 4/19/2023 1:07:45 PM
          Mass 69.00, Limits 2000000 to 6000000
                 EMV,      mass, abundance, peak width
                2678,     69.09,     169955,     10.80
                2878,     69.00,     426941,    -16.58
                3000,     69.17,     759604,      5.63
                3000,     67.92,     614101,      1.69
                3000,     69.04,     625889,      9.87
                3000,     69.07,     748013,      7.78
                3000,     68.89,     711960,      5.64
                3000,     69.15,     601906,     10.55
                3000,     68.98,     706757,      9.19
                3000,     69.00,     729832,      9.49
                3000,     69.19,     686079,     13.95
                3000,     72.14,     604682,     11.96
                3000,     72.70,     645750,     12.73
                3000,     68.84,     689163,      8.97
                3000,     68.97,     681005,      9.99
    Check Abundance failed
    Check Abundance, Quad 1 - Finished
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:09 PM
          target low mass 69.00, target high mass 502.00, tolerance 1.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
                0.000,      0.00,      1.00,   0.00048,     70.88,  709723.3,     10.66,    490.14,  672599.1,     11.51
    Adjust Mass Axis, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:08:13 PM
          target peak width 6.00, tolerance 2.00, adjustAB False
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -2.000,     25.00,   -0.0127,     20.87,     -2.40,   0.00060,     69.54,    707630,    10.756,    520.63,    633053,    49.137
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:15 PM
          target low mass 69.00, target high mass 502.00, tolerance 1.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
                0.000,      0.00,      1.00,   0.00048,     70.89,  773630.3,     20.76,    490.31,  635417.2,     49.89
    Adjust Mass Axis, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:08:17 PM
          target peak width 6.00, tolerance 2.00, adjustAB False
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -2.000,     25.00,   -0.5797,     -4.13,     -2.40,   0.00060,     69.60,    769411,    10.045,    499.80,    689953,    51.023
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:19 PM
          target low mass 69.00, target high mass 502.00, tolerance 1.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -2.600,     31.71,      1.00,   0.00048,     68.95,  916784.3,      4.22,    499.22,  629934.3,     33.04
    Adjust Mass Axis, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:08:22 PM
          target peak width 6.00, tolerance 2.00, adjustAB False
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -2.600,     44.88,   -1.3000,    -29.13,     -2.40,   0.00060,     69.88,    935183,     3.111,    515.38,    656675,     9.729
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:24 PM
          target low mass 69.00, target high mass 502.00, tolerance 1.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -1.492,     19.88,      1.00,   0.00048,     70.21,  659646.6,      8.76,    511.01,  684082.9,     41.76
    Adjust Mass Axis, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:08:26 PM
          target peak width 6.00, tolerance 2.00, adjustAB False
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -1.465,     -5.12,   -1.3000,    -54.13,     -2.40,   0.00060,     72.97,    643000,     8.954,    482.45,    643459,   132.719
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:28 PM
          target low mass 69.00, target high mass 502.00, tolerance 1.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -2.600,     19.88,      1.00,   0.00048,     69.78,  657049.3,     12.21,    524.33,  686829.1,     49.69
    Adjust Mass Axis, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:08:31 PM
          target peak width 6.00, tolerance 2.00, adjustAB False
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
               -0.600,     -5.12,   -1.3000,    -79.13,     -2.40,   0.00060,     64.88,    628196,    10.414,    523.72,    697160,    28.445
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Check Abundance, Quad 1 - Starting - 4/19/2023 1:08:33 PM
          Mass 69.00, Limits 2000000 to 6000000
                 EMV,      mass, abundance, peak width
                3000,     73.22,     647908,      9.44
                3000,     72.83,     602247,      9.80
                3000,     71.23,     669577,      3.50
                3000,     67.79,     628847,     10.02
                3000,     67.08,     610131,      8.88
                3000,     71.47,     601765,      9.54
                3000,     73.65,     599042,     12.29
                3000,     70.79,     601895,     13.08
                3000,     74.00,     673450,     11.38
                3000,     67.87,     638831,     12.70
                3000,     69.07,     656465,     14.02
                3000,     67.64,     612510,      2.99
                3000,     69.38,     669817,      7.02
                3000,     71.00,     616264,      9.95
                3000,     71.21,     661775,      8.91
    Check Abundance failed
    Check Abundance, Quad 1 - Finished
    
    --------------------------------------------------------------------------------
    Adjust Mass Axis, Quad 1 - Starting - 4/19/2023 1:08:57 PM
          target low mass 69.00, target high mass 502.00, tolerance 6.00, adjustAB False
          axis offset, axis gain,    axis A,    axis B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
                1.400,    -30.12,      1.00,   0.00048,     72.00,  626382.9,      8.59,    504.20,  637716.3,      0.96
    Adjust Mass Axis, Quad 1 - Finished
    
    --------------------------------------------------------------------------------
    Check Abundance, Quad 1 - Starting - 4/19/2023 1:09:00 PM
          Mass 69.00, Limits 2000000 to 6000000
                 EMV,      mass, abundance, peak width
                3000,     71.66,     623318,      6.97
                3000,     67.39,     651209,     15.85
                3000,     72.77,     634542,    -78.06
                3000,     64.51,     633425,     10.68
                3000,     66.86,     633261,     10.25
                3000,     73.02,     664850,     26.57
                3000,     68.16,     652672,      6.18
                3000,     68.52,     619051,      7.82
                3000,     70.02,     620432,      9.97
                3000,     72.01,     647605,     10.56
                3000,     64.40,     672849,      9.69
                3000,     73.72,     593919,     10.99
                3000,     67.16,     671230,      9.48
                3000,     68.10,     627462,      6.24
                3000,     67.28,     658006,      2.65
    Check Abundance failed
    Check Abundance, Quad 1 - Finished
    
    --------------------------------------------------------------------------------
    Adjust Peak Width, Quad 1 - Starting - 4/19/2023 1:09:24 PM
          target peak width 2.00, tolerance 1.00, adjustAB True
          axis offset, axis gain, pw offset,   pw gain,      pw A,      pw B,  low mass, abundance, peakwidth, high mass, abundance, peakwidth
                1.400,    -30.12,   -0.6583,   -100.00,     -2.40,   0.00060,     69.98,    629123,    12.329,    483.38,    681675,     4.721
               -0.600,     -5.12,    1.3000,    -75.00,     -2.40,   0.00060,     71.28,    637958,    36.463,    490.57,    663562,    17.997
               -2.600,     19.88,    1.3000,    -50.00,     -4.80,   0.00060,     68.86,    586803,    10.687,    516.90,    659104,    50.508
               -0.600,     -5.12,    1.3000,    -75.00,     -4.80,   0.00120,     69.73,    664196,     9.314,    502.12,    698250,    42.615
               -1.429,     -8.95,    1.3000,   -100.00,     -4.80,   0.00060,     68.74,    729573,     9.654,    495.94,    632753,    51.439
               -2.094,     16.05,    1.3000,   -100.00,     -4.80,   0.00075,     69.97,    669154,    10.156,    520.99,    648322,    34.612
               -0.196,     -8.95,    1.3000,   -100.00,     -4.80,   0.00075,     67.38,    613448,     4.757,    518.50,    678198,    31.849
                1.804,    -33.95,    0.8666,   -100.00,     -4.80,   0.00075,     72.30,    671255,    10.888,    523.13,    690330,    17.491
                1.545,    -58.95,    1.3000,   -100.00,     -7.20,   0.00075,     66.66,    694267,     9.795,    487.68,    659324,    50.752
                1.976,    -33.95,    1.3000,   -100.00,    -10.80,   0.00075,     74.00,    646836,     9.269,    490.36,    677181,    49.726
               -0.024,     -8.95,    1.3000,   -100.00,    -10.80,   0.00075,     73.39,    696147,     6.663,    501.40,    731119,    21.714
               -2.024,     15.03,    1.3000,   -100.00,    -10.80,   0.00075,     68.48,    679967,    10.890,    510.15,    656808,    49.909
               -0.117,     -9.97,    1.3000,   -100.00,    -10.80,   0.00075,     70.87,    672787,    -3.674,    480.69,    655374,    48.728
               -2.117,     15.03,   -0.4314,   -100.00,    -10.80,   0.00075,     69.98,    628689,    11.014,    487.72,    672352,    37.869
               -2.600,     40.03,    0.0686,   -100.00,    -10.83,   0.00075,     69.59,    662151,     9.442,    501.66,    652651,    47.016
               -2.600,     44.53,    0.2502,   -100.00,     -6.56,   0.00075,     68.31,    628944,    10.003,    505.40,    684482,    48.150
               -1.114,     24.84,    0.5705,   -100.00,     -4.01,   0.00075,     71.97,    652455,     8.480,    487.20,    632725,    49.527
               -2.600,     49.84,    0.5605,   -100.00,     -6.20,   0.00075,     66.84,    675531,    10.073,    508.92,    707443,    20.356
               -0.600,     24.84,    1.2631,   -100.00,    -12.21,   0.00075,     71.74,    642254,     8.750,    500.31,    729095,    11.612
               -2.600,     46.16,    1.3000,   -100.00,     -4.58,   0.00075,     74.00,    675602,     8.935,    482.15,    693998,     7.277
    Adjust Peak Width, Quad 1 - failed
    
    --------------------------------------------------------------------------------
    Adjust Peak Width failed
    

  • When was the last time the ion source was cleaned?  Did you try filament 2?  New filaments?  Is there PFTBA in the vial?

    It is setting the EM voltage to 3000V, the maximum, to get even 600k abundance. You need abundance to get the system to go further. Resolve that first and the peak width might be OK.

  • I have no idea when the source was last cleaned, it might have been a while ago (more than 1 year). Yes we tried filament 2 and the same error popped up. We checked the PFTBA vial and there are sufficient amounts left. 

    I will let my supervisor know tomorrow and then we might clean the source. I will also reply to this thread if the issue has been resolved.

    Thank you very much for your time and effort Paul, I really do appreciate it. 

Reply
  • I have no idea when the source was last cleaned, it might have been a while ago (more than 1 year). Yes we tried filament 2 and the same error popped up. We checked the PFTBA vial and there are sufficient amounts left. 

    I will let my supervisor know tomorrow and then we might clean the source. I will also reply to this thread if the issue has been resolved.

    Thank you very much for your time and effort Paul, I really do appreciate it. 

Children
Was this helpful?