I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
why Agilent didnt explain any thing about RFPA IN ANY MANUAL
i checked the highlighted cable and the area with necked eye and as i said above i changed the side and main board.
we already called local agilent support and they advise us to make new configuration for the instrument ; we did that but still the problem
another notice i saw it yesterday and have a vedio for it but did not know how to attach it that showing during the error of the instrument both the heat degree of ion source and quad are decreased about one to five degree and the vacuum gauge give a high vacuum reading and a message checking vacuum gauge then the instrument pass the error (taking about 90 second) after this the heat degree and vacuum return to normal value.
Nice video, u made..
Still RFPA problem coming..
Do tune evaluation and share..
If tune evaluation ok..then do autotune and once error come then take picture of that one and share to understand better..
Even, if possible then i like to see picture of this ms of that area where i suggest you to check..
tune evaluation is ok (pic down) and the error message appear (pic down)
Surprising, tune evaluation passed but autotune not..
dear ISH
during autotune the instrument generate report of autotune and u can inspect it or print the report . i printed it .the moment just you close the report , the instrument doing last step i don't know exact what it is but it include checking RFPA which was failed so it give the message tune action stopped RFPA FAULT
Here repeller value is showing 9.49..Even new system never show like this..
I don't know why but still i have doubt that there is some small things, not a big things...
Here repeller value is showing 9.49..Even new system never show like this..
I don't know why but still i have doubt that there is some small things, not a big things...
And Ion Body is zero in this report..
When u contact to your local agilent support then what they said abt these things...