I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
I HAVE GCMSHES 5977B ; during TUNING ;IT GIVES tune action stopped :MSD Fault 2 : RFPA Difficulty . WE try cleaning the ion source , changing filament but still the message appear
how can i resolve this?
why Agilent didnt explain any thing about RFPA IN ANY MANUAL
why Agilent didnt explain any thing about RFPA IN ANY MANUAL
Because that is service emgineer job to check RF power amplifier...
dear Mr.Well,
Still, i will suggest to check all wires from side board and ion source..
Quad frequency adjustment may also problem..
Otherwise it may electronics problem of power amplifier or side board..
Check around above pic that any wire is broken or bend, may be brown color..
i checked all wire and removed the black cover to see component of the board and all is ok
Then dipping of quad is only option.
And if still problem then onsite service call need to check electronics..