Hi, I have a GC 6890N with 5973N Mass selective detection, last week the computer failed and I had to change for another computer, on this computer windows XP, I installed the MSD ChemStation G1701EA E.02.00.493 software, on the  Agilent MSD configuration

Hi, I have a GC 6890N with 5973N Mass selective detection, last week the computer failed and I had to change for another computer, on this computer windows XP, I installed the MSD ChemStation G1701EA E.02.00.493 software, on the  Agilent MSD configuration I cheked the networking and the the instrument passed all Networks Tests, even I ping the IP of the mass detector and the GC and those pings are OK, but when I run MSD Chemstation a pop up window " MSINSCTL " shows " Mass Spect is already in use by other session". Somebody can help me, please.

attachments.zip
  • This problem can most often be fixed by restarting the smart card using the keypad. Using the keypad, press menu until you see the word controller, then press the item key until you see restart controller, then press yes. If you see reboot rejected, ChemStation present; then you will have to turn off the main power, wait 10 seconds, then turn it back on. At this point, I usually restart all other hw, GC, PC etc.

     

    If you are still running into the same error after restarting the MS, then the next thing to check would be the binding order of your network adapters. You want to make sure that the NIC you are using to communicate to the MS is at the top of the binding order. Here is a link to the Microsoft website that outlines how to update the binding order in Windows XP: https://support.microsoft.com/en-us/help/894564/how-to-change-the-binding-order-of-network-adapters-in-windows-xp-and

  • One thing I noticed on your first picture. Chemstation will not open if you have the Agilent configuration window open.

  • I marked this as assumed answered due to its age, if further help is needed, please post again and we will be happy to help

  • Can you change teh default gateway on the BootP to match 10.30.xxx.1?

    Also try changign the DNS to primary: 8.8.8.8, secondary 8.8.4.4

     

    you will have to reboot the system.

     

    in agilent configuration, can you click on the following and see if there is a connection:

    first the red to check if the instrument can actually be seen by the PC, if it passes, try the blue arrow and follow the instructions (note the blue one requires you to be physically in front of the instrument also ensure the power is not turned off and on until it tells you to).

  • Hi,

    I've been trying to set up two network cards as mentioned in the original post, but with no luck. I have a 6890A plus, and a 5973N, MSD Chemstation D.01.00 build 75 (G1701DA), loaded on a Dell Optiplex 330, which has an Intel CPU (so it's not an AMD thing in my case)

     

    I have tried setting static IP addresses for everything, so that our network doesn't try to use DHCP for anything (all of the addresses below are in the same subnet, but are outside of the range for the DHCP server). I have set up something like the following:

     

    Instrument NIC card

    10.30.xxx.20 - instrument connection (on-board NIC)

    255.255.255.0 - subnet mask

    10.30.xxx.1 - default gateway

     

    In BootP (with associated MAC addresses)

    10.30.xxx.21 - 6890N

    10.30.xxx.22 - 5073N

    10.30.xxx.20 - default gateway for both instruments

     

    Local Area Connection second PCI NIC card

    10.30.xxx.25

    255.255.255.0 - subnet mask

    10.30.xxx.1 - default gateway

    Added two DNS service for our network.

     

    MSD ChemStation still can't load the MS. 

     

    Anyone solved this?

  • I tried those things, and nothing has changed.  The error that came back from check network is a failed SC Query Test.  Just to be sure, I reinstalled the IO libraries.

     

    Any thoughts?

Was this helpful?