Hi
please create new instrument and configure with IP and check . for new instrument
go on control panel and instrument icon and create new instrument and configure with GC IP .
If issue not resolved repair the software and check .
Hi
please create new instrument and configure with IP and check . for new instrument
go on control panel and instrument icon and create new instrument and configure with GC IP .
If issue not resolved repair the software and check .
Dear Krishna
Thank you very much for the solution. It worked. But now the new instrument
does not have my old methods or data files. So every single time I have to
open methods and data files from target location? Is there a way to bring
all those files at one go?And are there any important parameters i have to
set before using as New Instrument?
Thank you for your time.