Hi
please create new instrument and configure with IP and check . for new instrument
go on control panel and instrument icon and create new instrument and configure with GC IP .
If issue not resolved repair the software and check .
Dear Krishna
Thank you very much for the solution. It worked. But now the new instrument
does not have my old methods or data files. So every single time I have to
open methods and data files from target location? Is there a way to bring
all those files at one go?And are there any important parameters i have to
set before using as New Instrument?
Thank you for your time.
sir
when you create new instrument and configure
in chem32 folder you will see two instrument will be there
1 and 2
1 is old and 2 is new
copy all method from 1 (method folder )and paste in 2 (method folder )
after that all method will come in your new instrument
same you can fellow with data and sequence folder
all method comes when you will launch the new instrument
Dear Krishna
Than you very much again. It worked!