modernization of the 7890 gas chromatograph

Hello !

The laboratory has a 7890 chromatograph with one evaporator and one FID detector. I want to upgrade it and add a second channel, the same, but for other tasks. I found catalog numbers on the Agilent website, but they are for the 7820 chromatograph, if I order G4328A (Agilent G4328A 0–100 psi Split Splitless EPC Inlet)  and G3462A (Agilent G3461A & G3462A Flame Ionization Detector (Packed and Capillary)), will it fit my chromatograph and the order will have a complete set of EPC module, electronics, tubes, evaporator with all the details? And everything will also be included for the detector - the EPC module, the detector itself, the electronics to it, the tubes? these modules from 7820 will fit to 7890? Thanks,

source link: https://www.agilent.com/search/?Ntt=G4328A 

                    https://www.agilent.com/search/?Ntt=G3462A&redirect=0 

BR Azrael

Parents Reply
  • The service manual is Agilent internal.

    PN G4328A listed above is for a 7820 GC, not 7890 - it will not be compatible.  For split/splitless inlet you'll need one of the following:

    • G3452A 100 psi split/splitless with EPC, Kit
    • G3460A 150 psi split/splitless inlet with EPC, Kit
    • G3453B Inert flow path, 0—100 psi (Same as G3452A but the top weldment and inlet body are inert-treated)

    For the FID you'll need one of:

    • G3461A FID with EPC, Kit adaptable to either packed or capillary columns.
    • G3462A Capillary FID with EPC

    Each kit should contain the full inlet or detector assembly including EPC.  I don't believe they can be ordered online - you'll need to contact your local sales representative.

Children
Was this helpful?