Recommended # of injections per vial cap/septum?

Hi all,

I am a novice in the practical aspects of GC (most of my background/knowledge is LC) so I realize this is a very basic question. After determining impurities in my samples were due to vial cap septum bleed, I was hoping to get some feedback on the following topics:

  1. Should vial caps be used for multiple injections in the same batch (i.e. blank injections)?
  2. If vials/ samples are reused, is it common practice to replace the caps after single use? Or can they be used again within a short time period (i.e. same day)?
  3. Is there a recommended vial septum material for ACN solvent? I know Agilent recommends rubber/butyl but I have seen conflicting information elsewhere (perhaps just difference in manufacturing?). 

I appreciate any and all help/ tips for a GC novice. 

Thanks,

Brian

Parents
  • Hi Brian,

    What kind of detector are you using? From my experience with FID, you shouldn’t have issues with septa bleed from re-running samples within the same day. If you inject the vial and then re-run it after a day, you will probably see n-hydrocarbon peaks related to septum bleed. Now this is with a FID, it may be more severe with GC-MS. Also, if you are using septa in your inlet, you’ll want to change that out about every 100 injections.

Reply
  • Hi Brian,

    What kind of detector are you using? From my experience with FID, you shouldn’t have issues with septa bleed from re-running samples within the same day. If you inject the vial and then re-run it after a day, you will probably see n-hydrocarbon peaks related to septum bleed. Now this is with a FID, it may be more severe with GC-MS. Also, if you are using septa in your inlet, you’ll want to change that out about every 100 injections.