GC ALS error 245

we are getting the ALS error 245 frequently ,ALS tower firmware also updated with A.11.04 .Even after Firmware version upgraded ALS error 245 has come two times.How to Minimize like this type of error.

Parents Reply
  • Hello  ,

    The light should flash a number of times, then stop, then repeat the cycle.

    Plunger errors are caused by:

    - Sticky syringes.

    - Improper syringe sizes.

    - Motor problems.

    - Mismatch between syringe configured in the SW and the physical syringe installed.

    - Viscous sample with improper plunger speeds.

    You can rule out most of those by simply removing the syringe and trying to run an injection. If it's still erroring out, it's a mechanical problem with the tower.