I m disturb with these message coming from my gc. 7890 a. ( see photos ). Can any one help me to fix it please ? I m sending also the picture of the manometer of the gaz
Dear all,
I m disturb with these message coming from my gc. 7890 a. ( see photos ). Can any one help me to fix it please ? I m sending also the picture of the manometer of the gaz
Hello rogermoise.
One possibility: if your are loading the method and the errors occur, it is possible you do not have enough total flow (low column flow, no or little split flow etc.) to the inlet to reach pressure before the inlet shutdown. If this is the case, load the method, manually increase the inlet total flow on the GC keypad until you reach pressure. Reload the method and all should be fine.
Hello rogermoise.
One possibility: if your are loading the method and the errors occur, it is possible you do not have enough total flow (low column flow, no or little split flow etc.) to the inlet to reach pressure before the inlet shutdown. If this is the case, load the method, manually increase the inlet total flow on the GC keypad until you reach pressure. Reload the method and all should be fine.