Front inlet pressure shut down

Dear all,

I m disturb with these message coming from my gc. 7890 a. ( see photos ). Can any one help me to fix it please ?  I m sending also the picture of the manometer of the gaz

attachments.zip
Parents
  • Hi,

     

    First, You should check below things:

     

    You can increase your helium pressure upto 100psi ( check picture 111713).

    Check column is okay. Means it's not broken.

    May be inlet is leaking so better replace septa.Check your inlet is properly sitting on top of gc.

    You should check column connection of inlet and detector.

    If you did not change vent filter from long time then better to change this.

Reply
  • Hi,

     

    First, You should check below things:

     

    You can increase your helium pressure upto 100psi ( check picture 111713).

    Check column is okay. Means it's not broken.

    May be inlet is leaking so better replace septa.Check your inlet is properly sitting on top of gc.

    You should check column connection of inlet and detector.

    If you did not change vent filter from long time then better to change this.

Children
No Data
Was this helpful?