I m disturb with these message coming from my gc. 7890 a. ( see photos ). Can any one help me to fix it please ? I m sending also the picture of the manometer of the gaz
Dear all,
I m disturb with these message coming from my gc. 7890 a. ( see photos ). Can any one help me to fix it please ? I m sending also the picture of the manometer of the gaz
You have an inlet leak. Have you done maintenance lately? If you haven't changed the septum in awhile it could be that or the nut may have loosened up and needs tightened a bit or the column nut may have loosened.