7890B Agilent GC Autosampler Issue

I have a 7890B agilent GC with an autosampler.

Recently a new error popped up. The error on the screen is #245 Autoinjector front tower plunger error.

Troubleshooting so far consists of the following:

 

I have inspected the old syringe and swapped it out with a new one. Problem persists.

I have inspected the unit, the plunger moves smoothly up and down by hand. The small black rubber belt looks in tact and free of damage.

The syringe is secured in place.

 

When I try to start a run, it goes to the wash cycle, once it drops into the wash vial and proceeds to draw up, the injector quickly retracts up and goes into error state.

 

How would I correct this issue?

 

We're beginning to think that a sensor error is occurring or a stepper motor error.

Alignment is next prior to sending in for repair.

 

Message was edited by: ***** Avalos

Parents
  • Hi javalos415,

    If you have tried a new ALS syringe, then the problem is most likely with the plunger motor itself. Replacement of the plunger motor would most likely be needed.

     

    Error codes are below:

    244(front), // EmptySyringe: syringe is empty but plunger home sensor not on

     

     

    245 (front), // FillSyringe: Plunger outside error band during filling. Error when more

    945(back) // than 140 counts behind or 70 counts ahead. (log_PlungerStuck)

    // EmptySyringe: Fast (< 10msec/%) mode and encoder above 1 count at

    // timeout. Slow (>= 10msec/%) and plunger falls 35 counts behind

    // or can’t reach 1 count when stopped. (log_PlungerStuck)

    //PartialEmpty: Plunger falls behind by 50 counts from desired trajectory

     

    Regards

    James

  • Hello.

     

    This is the exact error on the computer screen:

     

    Sampler error #245: Autoinjector front tower plunger erro.

Reply Children
Was this helpful?