Annoying regular ghost peaks?

Hi guys, I'm experiencing regular times ghost peaks starting at 120°C to 295°C.....my setup is GC-FID 6890, Low pressure drop, Ultra Inert Liner with glass wool, column HP1 widebore.....

 

I have cutted the inj/det 5 inches of column, heated at 295°C with 1mL/min Nitrogen flow, changed the liner...but nothing has changed.

 

The column is new, I only injected few samples of "ketones" in acetone.

 

The analysis below is a vial of HEXANE; I have those regular ghost peaks injecting acetone or acetonitrile as well.

 

What I have to do?

 

THANX

Parents
  • Ghost peaks that elute in a series like this are generally related to septum bleed. The regularly spaced peaks are indicative of a homologous series of siloxanes bleeding off of the silicone septa material. This could be coming from the septum itself or could be from septum chunks in the inlet and/or liner.

     

    What inlet septa material do you use? I recommend the Agilent Advanced Green or BTO septa.

     

    Have you tried changing the liner and gold inlet seal?

     

    Best regards,

    Jason

  • Thanx guys, I have

    -cleaned the inlet section with cotton swabs and isopropyl alcohol (body of injector/reducing nut)

    -replaced the split vent cartridge (it was like 5 years old..)

    -replaced liner and septum using advanced green septa

    -thermal heated the column then

     

    but...I found ferrules pieces between the gold seal and the liner!

     

    Now everything is good and my baseline is perfect.

     

    Thanx guys!

Reply
  • Thanx guys, I have

    -cleaned the inlet section with cotton swabs and isopropyl alcohol (body of injector/reducing nut)

    -replaced the split vent cartridge (it was like 5 years old..)

    -replaced liner and septum using advanced green septa

    -thermal heated the column then

     

    but...I found ferrules pieces between the gold seal and the liner!

     

    Now everything is good and my baseline is perfect.

     

    Thanx guys!