Troubleshooting front inlet flow problems

I require technical support for a 6890 GC.  When running a method with a temperature ramp, at the higher temperatures there is an inlet flow shutdown and the GC shuts down.  The upstream carrier pressure is 95 psi which should be high enough.  I notice, when setting up a method, that when I change the pressure for the back inlet, the front inlet pressure changes and may not reach the right setpoint depending on the pressure set for the back inlet.

 

I tried an "unobtrusive" leak test that I saw on some troubleshooting guide, which suggested going to a splitless flow and seeing if the flowrate in the front inlet was higher than the setpoint which would indicate a flow -- when I followed this procedure the front inlet had zero flow.

 

I'm wondering since the pressure from the back and front inlets aren't independent and I was not able to get front inlet flow from this test, if that means there is an issue with the flow module, or if anyone has any other ideas. 

Parents
  • Hi mj.hazlett,

    Can you tell me a little more about your inlet parameters? Column dimensions and approximate temp that the inlet shuts down would be useful. Oven ramp rate would be beneficial as well.

    Usually a shut down at higher oven temps indicates that you may have marginal total flow set for the inlet. Split/splitless inlets are designed to work with a minimum of 20ml/min. You can sometimes get away with lower total flow with isothermal or small rates of temp increase.  You are correct that one flow channel (inlet) shouldn't affect the other as they are on separate EPC modules. A supply of 95 psi should be fine as long as its 95psi to the back of the instrument and not meeting a restricted gas trap that would lower the pressure under flow conditions. 

  • The parameters are below; the flow issues start when the oven reaches 200C where the flow is half the setpoint, and the shut down occurs shortly after when the oven is around 220C.

     

    OVEN
       Initial temp:  50 'C (On)               Maximum temp:  300 'C
       Initial time:  4.00 min                 Equilibration time:  0.50 min
       Ramps:
          #  Rate  Final temp  Final time
          1 20.00      250        6.00
          2   0.0(Off)
       Post temp:  0 'C
       Post time:  0.00 min
       Run time:  20.00 min


    FRONT INLET (SPLIT/SPLITLESS)           BACK INLET (SIM DIST)
       Mode:  Split                            Pressure:  15.00 psi (On)
       Initial temp:  260 'C (On)              Gas type:  Nitrogen
       Pressure:  10.00 psi (On)
       Split ratio:  50:1
       Split flow:  115.7 mL/min
       Total flow:  120.5 mL/min
       Gas saver:  Off
       Gas type:  Nitrogen


    COLUMN 1                                COLUMN 2
       Capillary Column                        Capillary Column
       Model Number:  Agilent 19091J-413       Model Number:  Agilent 19091P-Q04
       HP-5  5% Phenyl Methyl Siloxane         HP-PLOT Q
       Max temperature:  325 'C                Max temperature:  290 'C
       Nominal length:  30.0 m                 Nominal length:  30.0 m
       Nominal diameter:  320.00 um            Nominal diameter:  320.00 um
       Nominal film thickness:  0.25 um        Nominal film thickness:  20.00 um
       Mode:  constant pressure                Mode:  constant pressure
       Pressure:  10.00 psi                    Pressure:  15.00 psi
       Nominal initial flow:  2.3 mL/min       Nominal initial flow:  2.3 mL/min
       Average velocity:  38 cm/sec            Average velocity:  43 cm/sec
       Inlet:  Front Inlet                     Inlet:  Back Inlet
       Outlet:  Front Detector                 Outlet:  Back Detector
       Outlet pressure:  ambient               Outlet pressure:  ambient


    FRONT DETECTOR (FID)                    BACK DETECTOR (TCD)
       Temperature:  270 'C (On)               Temperature:  250 'C (On)
       Hydrogen flow:  40.0 mL/min (On)        Reference flow:  20.0 mL/min (On)
       Air flow:  450.0 mL/min (On)            Mode:  Constant column+makeup flow
       Mode:  Constant makeup flow             Combined flow:  7.0 mL/min
       Makeup flow:  45.0 mL/min (On)          Makeup flow:  On
       Makeup Gas Type: Nitrogen               Makeup Gas Type: Nitrogen
       Flame:  On                              Filament:  On
       Electrometer:  On                       Negative polarity:  Off
       Lit offset:  2.0

  • I had a look at this photo and this is a custom analyzer rather than a basic GC.  Reconfiguration of what was an expensive custom analyzer goes beyond the scope of the Agilent community. We were able to find the original configuration and this was special order. I would not modify this GC for simple applications. I also see evidence that it may have been modified as the inlet labelling has been changed possibly by another company. If you wish to modify I would suggest contacting your local service Agilent | Contact Us

  • James, here is a photo of inside the oven.

     

    I ordered the filter and I am trying to clean out the split vent, honestly I am new to the instrument and finding the instructions on the service disk very confusing as I don't know which is the split vent line or where the trap is.....  Is there some schematic or picture of this you can send me?

     

    Sent from my iPhone

  • You can certainly try, I just dont know how the valves and plumbing will play into this. With some of these analyzers the back inlet was meant to only provide flow rather than be an active part of the injection path. There maybe someone else who is following this thread that maybe able to  add value.

  • Hello

     

    Can you elaborate more on this configuration?  So I want it to be single inlet to single detector for the purposes of these experiments... I have one liquid sample I want to analyze on front column with FID and gas sample on back column with TCD.  Is there a way I can make it like that?

     

    The application for liquid injection is that I want to get concentration of methanol in methanol water solution... I have been diluting this with acetone for analysis.  I was using kind of a blend between ASTM method D7768 and the Agilent Alcohols I method.

     

    I washed out the copper split vent line like you suggested.  For the trap, while I wait for the new one, should I rinse that as well?

     

     

     

    Sent from my iPhone

  • Hi James,

     

     

    I was previously unfamiliar with simulated distillation.  Can you elaborate on the issues of using the GC for applications other than this and why you might recommend reconfiguring?

  • Yes this GC has not been used in awhile.  There are two switching valves on the top (photo attached) and the serial number is US10451032.

     

    Sent from my iPhone

Reply Children
Was this helpful?