LAN adapter part number

Hello everyone,

 

I have a LAN issue on a GC 6890N/1530N (GC keeps going offline randomly and ping works randomly). I suspect it is from the LAN adaptater after changing ethernet cable and checking power saving features as described here.

 

I have an old GC, same model that has the same LAN adapter part number G1530-61910 (see attached picture).

However on Agilent website, the replacement kit part number is G1530-62010 as found here.

 

Is it the same ? Why is the part number different ?

 

Thanks in advance !

  • I assume PC is 10.1.1.100? I would make sure the hub or switch you are using is not the cause by swapping another in place. 

    The battery on the mainboard is there to hold parameters such as serial number etc when the GC is turned off and unplugged. Unless you are seeing loss of GC settings then the battery is ok. 

     

    Regards

    James 

  • If we can assume that all networking is ok. Please review below even though you confirmed.

    PC 10.1.1.100 , GC 10.1.1.101, MSD 10.1.1.102  , Subnet mask for all three 255.255.255.0

     

    There should be no gateway entry configured on the GC,MSD or PC, for a simple set up with just a PC and switch. 

     

    You could try a crossover cable direct from PC to the GC and test that way (GC only Configuration) . If you are still having the same issue, I would be looking at replacing the mainboard in the GC. See disclaimer below.

     

    The advice I give here is the best I can offer without being in front of the instrument. For replacing circuit boards and anything internal to the instrument I always recommend Agilent Trained Service personnel as electronics parts are not easily returned once opened. There are also matters of instrument and user safety , firmware, adjustments etc to be considered. The instruments are not meant to be user serviceable under the covers.

     

    Regards

    James

  • IP configuration is fine.

    I'll try to find a crossover cable to directly connect PC and GC and keep you posted.

    Thanks a lot for your help !!

    Regards

  • Hi James,

    Are these time delay of response normal for a GC ? Because for the MS I have systematically <1ms...

     

    Thanks for your help

    Regards

  • There are some notifications/warnings that may pop up on the GC display. These are from the manuals:

    For non-GPIB network connections as of early 1998, the 6890A would have used an HP JetDirect LAN card (like J2552B)

     The 6890N, LAN is built-in.

    Remember that these are very slow connections compared to what we are used to now, but they were state-of-the art in 1998!

     

    Warning 103—Signal 1 buffer full

    Warning 104—Signal 2 buffer full

    Status message: Signal 1 buffer full or Signal 2 buffer full

    This error could occur if you are using a data collection device, such as an

    integrator or ChemStation. If the data collection device goes off line during

    a run, signal data will be stored in the 6890 GC signal buffer. The buffer holds

    400 Kilobyte of data; if the amount of data exceeds the buffer’s storage limit,

    this warning will appear. No new data is stored after the buffer is full.

    ❐ There is a problem with the PC, the cabling to the PC, or the local network

    that links the GC to the PC. Check the PC, cabling, and network.

    ❐ The PC was turned off without closing the ChemStation instrument

    session. The GC collects and stores real-time plot data until the buffer

    overflows and the warning appears. Next time, close the instrument

    session before turning off the PC so that the GC stops collecting data.

    ❐ The PC entered power saver mode. When the PC enters power saver mode,

    its processor slows down and cannot collect data fast enough for normal

    communications, eventually causing the warning to appear. If the PC

    stays in power saver mode overnight, for example, there will be an error

    on the GC but the ChemStation will show a Ready status. Close and

    restart the instrument session, and disable the PC’s power saver feature.

    ❐ There was a software problem on the PC that stops data collection.

    ❐ There is a hardware problem in the GC. If the problem persists, contact

    Agilent Technologies for service.

     

    Warning 122—Host communication: data overrun

    Status message: Comm data overrun

    The 6890 GC received data from a computer or workstation faster than it

    could process it. The data may be lost

    ❐ Lower the baud rate or turn on handshake mode. You can adjust both of

    these parameters from the Communications control table, which is

    accessed from the Options control table.

    Warning 123—Host communication: data error

    Status message: Comm data error

    The 6890 GC received bad data from a PC or workstation. The data may be lost.

    ❐ The data bits or parity may be set incorrectly. You can adjust both of these

    parameters from the Communications control table, which is accessed

    from the Options control table.

     

    Warning 124—Host communications: abnormal break

    Status message: Comm abnormal break

    There is an interruption in communication between the 6890 GC and a PC or

    workstation. This could be caused by a LAN disturbance or a bad cable

    connection. Data may be lost

    ❐ If you are connected to the LAN, check to see if it is down or there is a

    problem with your connection to it.

    ❐ Check the cable connection between the GC and the computer or

    workstation.

     

     

Was this helpful?