When we start the Expert 2100 software, we get several pop up windows indicating an error that doesn´t allow the software to boot. The software worked fine for years, reinstalling it did not help.

 

First message:

"Error 5 (Invalid procedure call or argument) in procedure UserControl_Initialize of User Control AgtAssayProperties"

 

Second message:

"AgtUICompDotPlot. Run-time error '5': Invalid procedure call or argument"

 

Third message:

"Error 0 () in procedure IAgtAppContext_Update of Class Module CApp Context"

 

Is anyone able to help with these errors?

 

Thanks,

Tim

Parents
  • Hello,

     

     The 2100 Expert software error is being caused by a Microsoft Windows update that was released by Microsoft. Since the initial release, Microsoft has released optional updates for the majority of current Windows versions. But you will have to install the correct update for your system manually.

     

    Determine you Version of Windows

    1. From the start menu type Run and open the Run application.
    2. In the run application, please type winver and press enter. A windows will open with the necessary information.
    3. Note if it Windows 7 or 10 and not the Windows version.

     

     

    Determine your computer architecture

    1. Open the Start menu and type System Information
    2. Open the System Information Application and note the System Type.

     

     

    Download and Install the Update

    1. Navigate to the website: http://www.catalog.update.microsoft.com/home.aspx
    2. Search for the required update for your Windows version. See the attachment for which update you need to install.
    3. Download the correct update for your computer architecture.
    4. Install the update following the on-screen prompts.
    5. Restart the computer.

     

    Best Regards,

    Ian Jaworski

    Electrophoresis Support

    Diagnostics and Genomics Group

    Agilent Technologies, Inc.

    electrophoresis@Agilent.com

     

Reply
  • Hello,

     

     The 2100 Expert software error is being caused by a Microsoft Windows update that was released by Microsoft. Since the initial release, Microsoft has released optional updates for the majority of current Windows versions. But you will have to install the correct update for your system manually.

     

    Determine you Version of Windows

    1. From the start menu type Run and open the Run application.
    2. In the run application, please type winver and press enter. A windows will open with the necessary information.
    3. Note if it Windows 7 or 10 and not the Windows version.

     

     

    Determine your computer architecture

    1. Open the Start menu and type System Information
    2. Open the System Information Application and note the System Type.

     

     

    Download and Install the Update

    1. Navigate to the website: http://www.catalog.update.microsoft.com/home.aspx
    2. Search for the required update for your Windows version. See the attachment for which update you need to install.
    3. Download the correct update for your computer architecture.
    4. Install the update following the on-screen prompts.
    5. Restart the computer.

     