How can I get Mass Hunter 5.1 to finishing uploading files?

Hello,

I am having an issue where sometimes the connection to the database server stops and the files being uploaded halts but never picks up again. 

I will insert the error I see below:

because there is an audit trail on my software it is hard to navigate and our IT personnel have not had training on this software.

Is there anything I can try?

Parents
  • Hi, 

    assuming, your network connection is stable, this error most likely occurs, if you just shut down masshunter without closing the batch correctly (go to "File" and then "Close")

    Additionally you can check, if something is stuck in the File Upload Queue.

    Open the "File Upload Queue Tools" from the Taskbar Icon, and check if you can resubmit anything. Normally there should be nothing in this queue after a few moments.

      

Reply