Good day everyone,
I'm quite new to ICP-MS maintenance and I have a question regarding performance report in our ICP-MS Agilent 7700. We had very high RSD, I had done following steps :
- Cleaning the nebulizer and nebulizer chamber
- Replace the tubing
- Cleaning the plasma torch and bonnet by soaking it in Aquaregia
- Cleaning the sample and skimmer cone
- Nebulizer test and Shortage test at the lenses were okay
- The instrument was completely switched off (including the vacuum pump) for a week prior to this analysis
Our EM needs to be replaced soon, as you can see in the performance report. Do you know any tips, how to proceed with the measurement, or what other parameters I need to check in the meantime while waiting for the new EM to arrive? Thanks a lot in advance !