Hello Scientists, can someone tell me how the ICP Expert II adjusts the intensities of the standard and sample with an internal standard ratio as the method has defined the intensity corrected formula? Also once the internal standard was set up, the intens

The method is to quantify Zn and Ti using Y as an internal standard. Calculate the corrected average intensity. See below. Is the same formula that ICP Expert II uses? have the highlighted intensities been corrected?

  • Hello Tina,
     
    Thank you for your quick response. Your answer is really helpful.
     
    I understand that adding the internal standard to the standard and sample simultaneously is to correct any changes that may affect the intensities of sample and standard. In ICP-OES we set up the ratio of internal standard in blank to 1:00. See below:
     
     
    The ratio from result table shown is by this formula:
     
    Ratio = Iis (analyte)/ Iis (blank)
     
    When the ratio>1, the intensity of internal standard in sample/ standard is higher than that in blank, the intensity of sample/standard should be adjusted lower
     
    When the ratio<1, the intensity of internal standard in sample/ standard is lower than that in blank, the intensity of sample/ standard should be adjusted higher.
     
    Based on this principle, the formula for the adjustment should be:
     
    Corrected intensity of analyte = intensity of analyte / ratio
     
    It matches our method.
     
    However, as per your email, the software uses the formula :
     
    Ratio = Iis (blank)/ Iis (analyte)
     
    But adjustment is by: corrected concentration = C (analyte) * ratio. The ratio will be different from the above result table that we obtained, but the corrected intensity will be the same as above.
     
    Anyway, the key point is how the software adjusts the readings. If the same principle is followed, there is no impact on our results.
     
     
    Thank you again,
     
     
    Best Regards,
    Lily Zhong
    Health & Nutrition
    Health Science - Manager, Document Reviewer
    SGS Canada Inc.
    11-145 Konrad Crescent
    Markham, ON, L3R 9T9
     
    Phone:   +1 905-305-0998

    E-mail:    yanping.zhong@sgs.com
     
  • Hello Tina,
     
    Thank you for your quick response. Your answer is really helpful.
     
    I understand that adding the internal standard to the standard and sample simultaneously is to correct any changes that may affect the intensities of sample and standard. In ICP-OES we set up the ratio of internal standard in blank to 1:00. See below:
     
     
    The ratio from result table shown is by this formula:
     
    Ratio = Iis (analyte)/ Iis (blank)
     
    When the ratio>1, the intensity of internal standard in sample/ standard is higher than that in blank, the intensity of sample/standard should be adjusted lower
     
    When the ratio<1, the intensity of internal standard in sample/ standard is lower than that in blank, the intensity of sample/ standard should be adjusted higher.
     
    Based on this principle, the formula for the adjustment should be:
     
    Corrected intensity of analyte = intensity of analyte / ratio
     
    It matches our method.
     
    However, as per your email, the software uses the formula :
     
    Ratio = Iis (blank)/ Iis (analyte)
     
    But adjustment is by: corrected concentration = C (analyte) * ratio. The ratio will be different from the above result table that we obtained, but the corrected intensity will be the same as above.
     
    Anyway, the key point is how the software adjusts the readings. If the same principle is followed, there is no impact on our results.
     
     
    Thank you again,
     
     
    Best Regards,
    Lily Zhong