Can i mesaure Br concetration by ICP-MS 7900?
Hello,
Can i mesaure Br concetration by ICP-MS 7900?
Hi cell, I once tried to analyze for Br with our ICP-MS 7800, and I got really good calibration line for mass 79 and 81. However, my sample was a water leach sample without any other possible contaminants. Detection then was something like 1 ppb in the water. After this first experiment my client only wanted to focus on the Si content in the same samples so I skipped doing the Br content after that.
Br is volatile so I don't know if you can do it if you have to digest samples first!
Hi cell, I once tried to analyze for Br with our ICP-MS 7800, and I got really good calibration line for mass 79 and 81. However, my sample was a water leach sample without any other possible contaminants. Detection then was something like 1 ppb in the water. After this first experiment my client only wanted to focus on the Si content in the same samples so I skipped doing the Br content after that.
Br is volatile so I don't know if you can do it if you have to digest samples first!