Issues with Boron sensitivity on 5800 ICP-OES

Hello,

I run Boron with 10 other analytes on ICP-OES. As a commercial lab we run a large amount of samples. I have noticed that all the other analytes in the control (QC) remains relatively and accurate through the run +/- 10%.  On the other hand B recovery is accurate 90% at the beginning of the run and by 60 samples, it drops by 50%. By the end of a long run, it is usually off by 90%. Any ideas on how to tackle this.

Parents
  • Hello,

    You facing issue with low recovery it means sample (B analyte) might not stable (Solution chemistry) or intensity (Plasma issue) is decrease due to many reasons over many sample analysis.

    Can you run ISTD ? check if its within B selected wavelength (Atomic or Ionic line). 

    Hope the use of proper ISTD may help you in this ! Check and reply

Reply
  • Hello,

    You facing issue with low recovery it means sample (B analyte) might not stable (Solution chemistry) or intensity (Plasma issue) is decrease due to many reasons over many sample analysis.

    Can you run ISTD ? check if its within B selected wavelength (Atomic or Ionic line). 

    Hope the use of proper ISTD may help you in this ! Check and reply