ICP OES 5900 SVDV Water Cooling Issue

Hello, randomly the status light for water cooling will not turn green upon turning on the chiller. Per the instrument troubleshooting I removed the water cooling filter, cleaned and replaced it. I ran a chiller  performance check and all steps passed. I noticed on the dashboard the peltier won't cool below 26.4C. I also see that the camera water flow is 0 L/min. Does anyone have any insights or has experienced this problem before? 

Parents Reply
  • Hey Juliana, the root cause was algae build up which caused a short. I ended up calling Agilent and the part was under the 1 year warranty and service package. They replaced the water manifold assembly and no issues since. With the warranty/service package, it cost $0 to fix. Highly recommend a service package or extended warranty for these kind of issues as well as changing the chiller fluid every 4 months. 

Children