I would like your opinion and experience in ICP-MS, what is the causes that the sensibility is so different in different days?
It needs to be more than, 1000/1500/7500 right??
Hello julia.jsd
Which is the model ?7900 maybe? The specifications en No gas Mode for 7900 is 7Li>3000 cts, 89Y 15000and 205 Tl> 9000 cts
Sensitivity can change for multiple reason: vacuum, sample introductions, tune solution, cones.....
My first suggestions will be change the cones
Thanks
Regards
Hello julia.jsd
Which is the model ?7900 maybe? The specifications en No gas Mode for 7900 is 7Li>3000 cts, 89Y 15000and 205 Tl> 9000 cts
Sensitivity can change for multiple reason: vacuum, sample introductions, tune solution, cones.....
My first suggestions will be change the cones