Hello all,
I would like your opinion and experience in ICP-MS, what is the causes that the sensibility is so different in different days?
It needs to be more than, 1000/1500/7500 right??
- I dont change the argon cilinder
Hello all,
I would like your opinion and experience in ICP-MS, what is the causes that the sensibility is so different in different days?
It needs to be more than, 1000/1500/7500 right??
Hello julia.jsd
Which is the model ?7900 maybe? The specifications en No gas Mode for 7900 is 7Li>3000 cts, 89Y 15000and 205 Tl> 9000 cts
Sensitivity can change for multiple reason: vacuum, sample introductions, tune solution, cones.....
My first suggestions will be change the cones
Thanks
Regards