Hi,
I have an MP-AES 4100 that works with a Nitrogen generator. Is there any trouble shoot pdf document of it since its giving us some trouble and i would like to see how we can solve this issu
Find maintenance, best practice, and troubleshooting articles.
Hi,
I have an MP-AES 4100 that works with a Nitrogen generator. Is there any trouble shoot pdf document of it since its giving us some trouble and i would like to see how we can solve this issu
Hello Drinu
I have a similar instrument MP-AES 4210.. It would be more helpful to narrow the problem if there is any.. I do not think there's a trouble shooting PDF apart from the help software attached to the instrument.
Regards
Kalunga
Hi,
Basically, I think that there is some water trapped in it since the was disturbing the flow of the plasma and when we swabbed with generator with N cylinders the instrument worked fine with no errors. I am wondering if there is a flushout procedure that one can do
Regards
Andrew
Hi I've sent you a private message. When you have time can you please send me these steps
Bonjour Drinu, j'aimerais aussi profiter de ton expertise. J'ai un MPAES 4200,
Good Morning Hermann
Well, I have the MPAES-4100 and we have some trouble with the Nitrogen generator. Basically, we need to change the inlet filters due to the plasma is not stable due to the filters are blocked.
Nous n'avons pas de problème avec le générateur d'air. Notre problème se situe au niveau des résultats et de l'optimisation. Pour les résultats, Nous n'arrivons pas à tracer la courbe de calibration pour nombreux paramètres malgré les dispositions prises au niveau de la qualité des standards.
First, check that the standards that you are using fresh and are issued from the Agilent. Also, try to adjust the parameters of the metals and check also the nebulizer that is in the correct form. we had some issues with the Chromium and Zinc lately but when we adjusted their parameters the results came good as aspected
Merci pour ta disponibilité et pour ton conseil. Je travaille dans un laboratoire qui surveille la qualité de l'eau potable. Nous travaillons que sur l'eau. Pourrais je avoir quelques paramètres d'optimisation pour certains métaux ?
Merci pour ta disponibilité et pour ton conseil. Je travaille dans un laboratoire qui surveille la qualité de l'eau potable. Nous travaillons que sur l'eau. Pourrais je avoir quelques paramètres d'optimisation pour certains métaux ?
send me your email address via private message and will send you the ones that we use maybe they can help you.