Measuring Cr in high ferrous material

Hi
does anyone have an explanation for the fact, that I have a lower result for Cr (approx. 30%) with material with a high iron content, e.g. Millscale, with the measuring mode 'NoGas'? Measured with He mode, the results are correct. Measured with ICP-MS Agilent 7800, digestion in HNO3, HBF4 and water.
Thank’s and regards

Parents
  • Hi, thanks for the good input, I hadn't thought in this direction. The effect only occurs with materials that contain a lot of iron, in the middle of a batch that also contains other materials, so I thought it was an influence of Fe. No such effect occurs with other materials.

    I will have another look at this next week, best regards best regards

Reply
  • Hi, thanks for the good input, I hadn't thought in this direction. The effect only occurs with materials that contain a lot of iron, in the middle of a batch that also contains other materials, so I thought it was an influence of Fe. No such effect occurs with other materials.

    I will have another look at this next week, best regards best regards