ICP-MS 7700X abrupt shutdown.

Hi,
ICP-MS 7700X plasma shuts down abruptly after a while. The software cuts off communication without any error codes occurring. At the same time, the vacuum pump also turns off. We tried to fix the problem by inactivating the power button of the device, but the problem could not be solved. We connected the vacuum pump to another power source but again the problem was not resolved. please help with the problem!

Best regards.

Parents Reply
  • Hi , I'd like to inform you that our instrument problem was solved, and yes it caused by the broken UPS, the output voltage is dropped to 160 volts when igniting the plasma, and also the voltage was not stable at all and unfortunately, it also made our instrument power supply make a "boom" sound when switched on. But now everything is alright after we repaired the UPS and replaced the 48V power supply, hopefully the same with your ICPMS