Is it possible to upgrade/convert ICP-OES 720 series to ICP-MS by equipment add-ons? We run an analytical centre at University of Toronto/ChemEng Dept. and had users ask this question.
Find maintenance, best practice, and troubleshooting articles.
Is it possible to upgrade/convert ICP-OES 720 series to ICP-MS by equipment add-ons? We run an analytical centre at University of Toronto/ChemEng Dept. and had users ask this question.
Hello, unfortunately it is not possible to convert a 720 series ICP-OES into an ICP-MS. The same is true for any model ICP-OES. While the ICP source is similar it is specific for the detector (OES/MS) and they are not at all interchangeable.
If you have a need to replace ICP-OES with ICP-MS I would advise you to contact your Agilent sales representative and discuss trade-in options and against the new technology you require.
Hope this helps!
Kind regards,
Gareth
Thank you, Gareth. Your answer was very helpful indeed, especially the trade-in idea.
Best regards,
Georgiana
(Personal Details Removed by Moderator)
This question has been marked as assumed answered.