Due to helium shortage, we need to change carrier gas for our GC-FID. Most documentation for Agilent suggests either H2 or N2, and the Chemstation configuration shows those options as well.
We are considering using argon, however, I…
Carbon disulfide is another option for wash solvent when running naphtha or crude oil samples. It's response on FID is low compared to hydrocarbons, so there should be little to no response if some residual is in the syringe.
Since the FID and TCD are in series, I would guess that the extra noise on the peak through the FID is due to the TCD valve switching between carrier and reference. Try setting the FID signal rate to 5 Hz. Also, the FID makeup flow can be reduced or turned…
There's a lot going on. You have Column 1 in Constant Flow to a PSD in Constant Pressure with PSD Purge at 0 ml/min. Is there a purged splitter plate installed or a Purged Ultimate Union? The flow into the MS changes throughout the run due to the PSD…
Here is a collection of Agilent LC Resources. Links to user manuals, firmware resources and maintenance videos , application resources , Agilent InterAct! (Virtual LC Simulation) , HPLC Advisor App , LC Supplies and Selector Tools and LC Troubleshooting…
This problem occurred after updating Openlab CDS Software 2.2 Version to 2.4 Version. GC# 7890B With FID , and TCD Detectors With OpenLab CDS Software 2.4 After installing the software version 2.4 and after a gas sample injection I noticed that there…
Dear Colleagues,
I have GC-FID 8890 and i'm using Gas saver feature but I'm observing when I change time of gas saver , the retention time is changed.
for instance, I set gas saver at 20 ml/min after 2 min then inject methanol and the RT is 4.74 but…
The boxes of standards should have shipped with a Certificate of Analysis enclosed. If you go to Agilent.com, you can search on the part number. A number of the hits are CofAs from different batches. part 5080-8858 CofA Lot CF-4631W.pdf (agilent.com)…
Ah, and I think the issue has to do with that I have 2 autosamplers, and 3 columns (MS at back, and 2 different FID columns at front using a 2way splitter). So the MS signal is in the subsequent folder, the FID1 signal is in the correct folder, and FID2…
Hello,
To calculate the RSD and Pass/Fail results, one can try the following:
In the reporting go to the edtor and open "table properties"
In my specific case, I want to calculate RSD for Area, so I choose the column A, pA*s and press "Column…
This Information Applies To: Agilent 8890/8860 GC Systems
Issue
Early Maintenance Feedback is a useful feature on Agilent GC Systems and software that allows a user to monitor the maintenance needed on their GC. This feature allows for users to...
This Information Applies To: Agilent 8890/8860 GC Systems
Issue:
This procedure outlines the steps to view and clear the maintenance warnings on based on Agilent 8890/8860 GCs. You can also follow the onscreen instructions to perform maintenance...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with Agilent MS-40+ rotary vane pump.
Issue
Regular check of the rough pump fluid level in your MS40+ pump as part of regular 6000 LC/MS system maintenance checks.
Steps...
This Information Applies To the Following Products : All Agilent iFunnel and Agilent Jet Stream (AJS) 6400 Series LC/MS triple quadrupole systems, Agilent 6460A/C, 6470A, 6490A, and 6495A/B/C
Issue:
As LC Triple Quadrupole Mass Spectrometers (LC...