Hello Marty,
thanks for your reply. I tested your hint. It worked and I'm confident to code the formula in the CC.
Thank you very much and have a nice weekend,
Sarah
Hello well ,
I know this is an old forum post, but just wanted to add to what ISH was saying.
The answer still stands, 6890 is not compatible with the 7650 ALS. The GC models it is compatible with are (at the time of this comment at least):
- 8890…
Hi,
I passed this query on to a colleague who was involved with developing these columns, this was their response:
"It might be worth checking if they are filtering their mobile-phase and samples as it's possible the guard may be fouled. They could…
Welcome to the Dissolution Wiki.
Please find the most useful Dissolution resources from Agilent in the list below.
Dissolution catalog
Dissolution webinars
Complementary Dissolution 1-on-1 self-paced training
Contact Agilent dissolution expert…
Thank you for contacting us. Please note that analyzing the raw data file that ends in .xad is more informative than the screenshot shared. However, based on the screenshot shared, the ladder appears to have run as expected, therefore indicating a potential…
Hello rcbeck ,
I apologize for the delayed response.
There was a known issue with some of the older shipping templates that had an incorrect expression to evaluate the LevelEnable boolean. I would suggest using something like this and testing the…
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...
This Information Applies To: Agilent 7697A Headspace Sampler
Issue:
There is a potential leak external to the Headspace unit that can cause loss of pressure.
Steps to Follow:
This article is a troubleshooting guide for when you suspect there...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if an Agilent support representative...