Argon quality for the Agilent 700, 5100 and 5110 icp-oes systems is 5.0 or better. However the 5800/5900 will work with the 4.5 (99.995%) argon quality. For the optional oxygen addition, needed for some organic matrix measurements like kerosine a mix…
Hi, yes we did resolve the issue and it was a faulty Z-motor. Motor was replaced and "Z-axis" error went away. The "unexpected collision" error was due to me using the wrong vial bottoms with the vial tray I had. I needed to use rounded bottoms instead…
Hello Martin and Andy! I found out what were wrong. In my sequence two injections doesn't have a DA Method linked. When I pin only the injections with DA Method linked, the "No Data Found" was gone.
Thanks for letting us know. I will look into this and determine why it is not working. In the meantime, the file can be downloaded from the 7820A product page at https://www.agilent.com/en/product/gas-chromatography/gc-systems/7820a-gc-system . Scroll…
Thank you very much for your quick answer. It's very important for me to understand the concept well, so I have some consistency problems with the response of the compounds.
The Digest of upcoming webinars has been updated as of May 23.
The new Agilent Community Group, GPC/SEC Solutions Group hosted a webinar series in partnership with Dow in April: A Journey through SEC, Light Scattering, and 2D-LC for Macromolecules …
For more information, check out the application note Five Keys to Unlock Maximum Performance in the Analysis of Over 200 Pesticides in Challenging Food Matrices .
Recommended parts from the video:
Self-tightening, collared column nut for use with…
Hello,
Are you still having low flow issues on your AA? Did replacing the nebulizer assembly help?
Do you change the flame gases for your methods? If so, call up a default air/acetylene method. You will need a lamp of that element. Go to Optimize…
This Information Applies To: 1100, 1200 and 1260 Infinity I and 1260 Infinity II LC Pumps (Excluding preparative pumps with electromagnetic purge valve (EMPV): G1361A. Not applicable to any of the 1290 Infinity II modules with automatic purge valve: G4220X...
This Information Applies To: Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
Issue:
Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules.
Both the visible and UV lamps are consumable...
This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps.
Issue:
Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module.
Steps...
This Information Applies To: All Agilent 1260 Infinity Binary, Quaternary, or Isocratic Pumps without the Seal Wash Option installed.
Issue:
Worn pump seals can shed small particles that can cause blockages and pressure issues downstream of the...
This Information Applies To: Agilent G7167x Multisampler modules
Issue:
Required when:
The needle seat is visibly damaged, blocked, or is leaking.
Leaks or blockages in the Vialsampler are observed, especially between the needle and needle...
This Information Applies To: Agilent 1260 Isocratic (G1310B) Quaternary (G1311B) Pumps.
Issue
Required when:
The inlet valve is defective.
The pressure or leak test in Lab Advisor suggests the inlet valve may require replacement.
A large...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
Required when:
The outlet valve is defective or leaking.
The pressure or leak test in Lab Advisor suggests the outlet valve may require replacement...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
The seal wear in procedure is important as it acts to "bed-in" the seals to ensure a leak free and pressure-tight pump head.
Required when:...
This Information Applies To: Agilent G7167A/B Infinity II Multisamplers
Issue:
The Agilent Multisampler autoreferencing procedure uses predefined positions on the base plate and the sample hotel to calibrate the positioning of the needle parkstation...
This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
Issue:
Reinstallation of the pump head on the pump metering drive during maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump...