Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to Autoreference the G7167A/B Infinity II Multisampler
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Injectors and Sample Thermostats
  • Product Name :  1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity II Prime LC System 1290 Infinity II Bio LC System 1290 Infinity II LC System
  • Task :  Maintenance Operation
Related Articles
  • Troubleshooting Autoreferencing Issues in Multisamplers

    This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler Issue The needle bends while moving to the wash port or showing one the following errors in the acquisition software: EE 25034,0: 'Sampler alignment procedure command failed' EE 25098,0: 'Alignment procedure: hotel command failed' EE 25151,x: 'Axis did not reach its target position in time' EE 25121,0: 'Sampler...
  • How to Change the Needle Seat on a G7167x Multisampler

    This Information Applies To: Agilent G7167x Multisampler modules Issue: Required when: The needle seat is visibly damaged, blocked, or is leaking. Leaks or blockages in the Vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of the yearly maintenance procedures. Steps to follow Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Flathead screwdriver...
  • How to Change the Needle on a G7167x Multisampler

    This Information Applies To: Agilent G7167A/B Multisampler. Issue: Required when: The needle is visibly damaged. Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of yearly maintenance Steps to follow: Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Parts Required: G4267-87201 - Needle assembly OR G4267-87210 -...
  • Change Needle Warning Message on TapeStation Systems

    This Information Applies To: Agilent 4150 and 4200 TapeStation systems. Issue The TapeStation controller software displays a warning message, notify users to replace the needle after 3840 piercings. If the maximum number of 4480 pierces is reached, further analysis is inhibited. Exchanging the needle when the warning message appears ensures optimal piercing of the ScreenTape. Items Required Needle Change Cartridge (p/n 5067-5783 ) Steps to Follow Exchange the piercing needle ...
  • How to Set Up Bottom Sensing in Multisamplers

    This Information Applies To: Agilent G7167A/B, G5668A, G7137A, and G4767A Multisamplers Issue The bottom sensing feature in Agilent OpenLab Software is used to detect depths of vials or plates. Material of plates or vials must be able to resist the force of the needle; not all vials are suited. Items Required Compatible consumables for bottom sensing feature ( Figure 1 ): Glass vials without inserts (such as 5182-0714 ) Glass vials with fixed inserts (such as 5188-6591 ) Glass...

How to Autoreference the G7167A/B Infinity II Multisampler

Answer

This Information Applies To: Agilent G7167A/B Infinity II Multisamplers


Issue:

The Agilent Multisampler autoreferencing procedure uses predefined positions on the base plate and the sample hotel to calibrate the positioning of the needle parkstation and the sample hotel. Auto referencing is required to compensate for deviations in positioning the needle assembly and the sample tray. The procedure is required after disassembling the system or when you exchange the sample handler, the sample hotel, the needle parkstation, the needle assembly, or one of the main boards.

This function is implemented in the drawer detection and in the needle exchange routine.


Steps to follow

There are various ways to perform the auto referencing procedure of the G7167A/B multisampler (via Lab Advisor, CDS, or the Instant Pilot).

All the methods first require the following preparation steps:

  1. Empty the workspace of the Multisampler
  2. Close all drawers properly
  3. Install two sample trays in each drawer without sample containers
  4. All drawers should have been properly configured
  5. The needle assembly is installed in the needle parkstation

Auto Referencing via Lab Advisor

  1. Connect the instrument to Lab Advisor
  2. Go to Instrument Control
  3. Go to the Controls of the Multisampler and find Configuration
  4. Press the "Clear Data" button 
  5. Reboot the sampler with the on/off button
  6. Reconnect the Lab Advisor go back to Instrument Control and Configuration
  7. Click the Detect Drawers / Auto Referencing button and follow the procedure.

Auto Referencing via CDS

  1. Open the CDS of the instrument

  2. Right-click in the Active Area of the Multisampler. This action will open a drop-down menu. Select 'Modify' and select 'Drawer Configuration' (Figure 1).


    Figure 1. Drawer Configuration in OpenLab CDS


  3. Follow the steps given by the software for the Autoreferencing procedure

  4. When done, click the 'Back' button.

Auto Referencing via the Instant Pilot

  1. Connect the instrument with the Instant Pilot
  2. Press More and then press Diagnostics
  3. Find the tests under the Multisampler. Perform the steps for the Autoreferencing procedure given by Instant Pilot.

 

Tip
Learn how to effectively maintain, troubleshoot, and operate your Agilent LC System:
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning path available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help