Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System
  • Task :  Maintenance
Related Articles
  • How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump

    This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules Issue: Required when: The outlet valve is defective or leaking. The pressure or leak test in Lab Advisor suggests the outlet valve may require replacement. There is large increase in pressure ripple is observed. Steps to follow Tools Required: 8710–1924 - Wrench, open 14 mm 8710–0510 - Wrench, open 1/4 - 5/16 inch 5067–5688 - Torque wrench, 1 - 25 Nm with 14 mm wrench Parts...
  • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX

    This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps Issue: How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps. Resolution: The following information should help you find what part of the pump is causing a flow or pressure problem. Figure 1 and Figure 2 show the hydraulic paths of the pumps. Figure 1 - Hydraulic path for the Isocratic and quaternary pump...
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules Issue: Reinstallation of the pump head on the pump metering drive during maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module Steps to follow 1. Ensure that your Agilent 1260 Infinity Pump module is turned off at the front power switch. 2. Place your pump head on the pump metering drive. 3. Insert the two pump head bolts through the two holes in the pump head...
  • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps. Issue: Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module. Steps to follow Caution : When loosening the pump head blots, make sure that the bolts are loosening evenly left and right. Do not loosen one pump bolt fully while leaving the other bolt tight as it may cause misalignment of the pump head...
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option

    This Information Applies To: All Agilent 1260 Infinity Binary, Quaternary, or Isocratic Pumps without the Seal Wash Option installed. Issue: Worn pump seals can shed small particles that can cause blockages and pressure issues downstream of the seals. Agilent recommends that pump seals are changed yearly, or if pump performance indicates seal wear. Indications of seal wear include leaks at the lower pump head side, unstable retention times, or unstable pressure ripple. Replacement of the...

Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps

Answer

This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules 

 

Issue:

The seal wear in procedure is important as it acts to "bed-in" the seals to ensure a leak free and pressure-tight pump head.

 

Required when:

  • The PTFE (black) piston seals have been replaced in the pump head.
    See: How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option

 

Tools Required:

  • 0100-1847 - Adapter AIV to solvent inlet tubes
  • 5022-2159 - Restriction capillary
  • LC-grade isopropanol

 

Parts Required:

  • 5063-6589 - Piston seal, PTFE (2/pk)

 

Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents.  When working with these substances observe safety procedures (for example by wearing protective goggles, suitable safety gloves, and protective clothing) as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.

 

Caution: Seal Damage - This procedure is required for black PTFE seals (for standard applications p/n 5063-6589), but it will damage the yellow PE seals (for normal phase applications p/n 0905-1420). Do not run this seal wear-in procedure if PE seals are installed in the pump head.

 

Before replacing your solvent with isopropanol or replacing isopropanol with your solvent, consider solvent miscibility. For example, do not directly switch from buffers to isopropanol and vice versa.

 

1. Place a bottle with 100 mL of isopropanol in the solvent cabinet and put a tubing (including bottle head assembly) into the bottle.

 

2. If an active inlet valve (AIV) is installed, screw the PEEK adapter (0100-1847) to the AIV and connect the inlet tube from the bottle head directly to it. If a Passive Inlet valve (PIV) is installed then connect the solvent line directly to the PIV. See Figure 1 for an illustration of each valve.

  

 

Figure 1. Active Inlet Valve (left) and Passive Inlet Valve (right)

 

3. Connect the Restriction Capillary (5022-2159) to the purge valve. Connect the other end of the capillary to a suitable waste container.

 

4. Open the purge valve and purge the system for 5 mins with LC-grade isopropanol at a flow rate of 2 ml/min.

 

5. Close the purge valve and set the flow to a rate adequate to achieve a pressure of 350 bar. Pump for 15 mins at this pressure to wear in the seals. The pressure can be monitored using your instrument control software.

 

6. Turn OFF the pump, slowly open the purge valve to release pressure from the system.

 

7. Disconnect the restriction capillary and reinstall the bottle with the solvent for your application.

 

8. Rinse your system with the solvent used for your next application.

 

9. Replace the purge valve frit. See: How to replace the PTFE frit of the Purge Valve in LC Pumps

 

 

 