Hello Elei , This is just the normal system pressure as measured by the pressure sensor inside the pump. You can check this from the online or acquisition section of your SW, under the pump status. What SW are you currently using?
Hi MSVS , My pleasure to assist. Thank you for letting us know that you were able to configure successfully using the steps. I am glad to hear that we now have the solution to this case.
The Digest of upcoming webinars has been updated as of May 29.
Notable additions include:
GC & GCMS (1)
OMICS (1)
Check Out:
EPA 200.8 Water ICP-MS Analyzer – Learn How This Integrated Package Ensures Your Lab is Up and Running Samples…
Soap and water or Snoop type leak detection fluids are not recommended to be used on GC or GC/MS gas connections. If gas is leaking out, gas is also leaking in - that's why a leak in the helium line will show as air in the background. That means that…
Here is a "best of" collection of links to user manuals and resources for OpenLab CDS on Agilent.com.
For OpenLab ChemStation, see Collection of OpenLab ChemStation Resources .
For OpenLab EZChrom, see Collection of OpenLab EZChrom Resources . …
Hello Adamcfc101 Please advise if the intermittent error "metering device command failed, draw command aborted" is still occurring on your dual needle multisampler. As Carlos has asked, please also share the following details:
- is this error happening…
Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
This Information Applies To: 1100, 1200 and 1260 Infinity I and 1260 Infinity II LC Pumps (Excluding preparative pumps with electromagnetic purge valve (EMPV): G1361A. Not applicable to any of the 1290 Infinity II modules with automatic purge valve: G4220X...
This Information Applies To: Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
Issue:
Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules.
Both the visible and UV lamps are consumable...
This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps.
Issue:
Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module.
Steps...
This Information Applies To: All Agilent 1260 Infinity Binary, Quaternary, or Isocratic Pumps without the Seal Wash Option installed.
Issue:
Worn pump seals can shed small particles that can cause blockages and pressure issues downstream of the...
This Information Applies To: Agilent G7167x Multisampler modules
Issue:
Required when:
The needle seat is visibly damaged, blocked, or is leaking.
Leaks or blockages in the Vialsampler are observed, especially between the needle and needle...
This Information Applies To: Agilent 1260 Isocratic (G1310B) Quaternary (G1311B) Pumps.
Issue
Required when:
The inlet valve is defective.
The pressure or leak test in Lab Advisor suggests the inlet valve may require replacement.
A large...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
Required when:
The outlet valve is defective or leaking.
The pressure or leak test in Lab Advisor suggests the outlet valve may require replacement...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
The seal wear in procedure is important as it acts to "bed-in" the seals to ensure a leak free and pressure-tight pump head.
Required when:...
This Information Applies To: Agilent G7167A/B Infinity II Multisamplers
Issue:
The Agilent Multisampler autoreferencing procedure uses predefined positions on the base plate and the sample hotel to calibrate the positioning of the needle parkstation...
This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
Issue:
Reinstallation of the pump head on the pump metering drive during maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump...