The Introduction Service if purchased includes the following:
Customer guidance to reconfigure the GC from checkout to intended use, including a single injection of a QC standard for the desired method. In the absence of a pre-existing method, the…
In your acquisition sequence table for your calibration samples, is this selected?
Do you use the same name for your data analysis method over time? Are there multiple copies of it on your system? Did you start a new batch from the beginning or add…
Carefully remove the old O ring. There's a spot in the lower right hand edge that has an opening where you can very carefully pry it out. Do not scratch the aluminum manifold.
Use a solvent-dipped cotton-tipped swab to wipe out the entire O ring channel…
Here is a "best of" collection of links to manuals, tools, and other resources for Vacuum and Leak Detection Solutions .
Products are listed on the Vacuum Technologies home page. Get there by going to products on Agilent.com, choosing Vacuum Technologies…
Hello Kris, this valve is normally open and if you put a flowmeter on this vent line you should measure what is shown on the inlet Split Flow setting (in split mode) or Purge Fow (in Splitless mode). If Gas Saver is enabled and active it should measure…
The Agilent Cary 3500 Flexible UV-Vis system is a versatile instrument used for analyzing a privacy screen's light transmission properties at various angles and light conditions. It boasts a spacious sample compartment suitable for different accessories…
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...