I guess checking the settings will be the first thing. On the GC side, since RT did not shift too much, I think oven temp would be of a less issue. But flow and pressure might act more in this case. Based on info I learnt from here, both pressure and…
Hello AJohnson13 ,
Lab Advisor can be currently downloaded from this page.
Lab Advisor Software Download | Agilent
This may change over time. You can search the web for 'agilent lab advisor download' to find a link to the current download location…
Hello,
So, you should be able to add another column to your report table and change the value to = Compound_Concentration * Sample_Amount / 100 to remove the sample amount and percentage. If you need to remove a multiplier or dilution you will need…
Hello,
LIMS ID is not available as a token for sample name, datafile name, or in CDS DA for report output. You can see the options in the screenshot below. It can be a value in the CSV file but not used in the name. The only other option would be…
Our G4220A (1290 Bin Pump) is setup to work with the seal wash currently. It operates with our instrument with no issue. If I go into the instrument configuration and check the box to enable the pump seal wash (The pump is installed), it will lose connection…
Hi RDP0314,
One thing to check is the injection settings. This behavior is common when the GC is expecting a series of manual injections. To check, click on the Instrument menu shown on your screen capture above and 'Select Injection Source' or similar…
This Information Applies To: Agilent OpenLab CDS
Issue
Identifying the baseline codes and baseline code descriptions for OpenLab CDS
Background
OpenLab CDS Data Analysis includes the same integration algorithms as available in both OpenLAB...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to filter a signal in an Intelligent Report.
Requirements
To be able to carry out the procedure as described, the privilege of Data Processing...
This Information Applies To: Agilent OpenLab CDS
Issue
This article outlines different ways to quantify the amount of compound in an unknown sample.
Background
Quantitation uses peak area or height to determine the amount of a compound in...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle seat of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers
Issue
Replacing the Needle Assembly of the Autosampler is required when:
The needle is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here
Issue...
This Information Applies To: Agilent 6000 series LC/MS instruments
Issue:
Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs.
Steps to follow ...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in...