Hello,
Having sorted out our printer problem with MassHunter for GC we started having another issue on Friday night last week. The software suddenly stopped mid run giving an error that the drivers for the MSD could not be found. After restarting…
Hello mr.white, You can do this with a sequence summary report template: Use 'Multi Signal Plot Overlaid' as you've done already; in the properties, the signals need to be overlaid. You may group by signal e.g. FID Front signal/ FID Back signal or, if…
Hi. The pump made loud noises for several days. After that, the pump stopped working normally. Near 600 rpm the pump stops and A-PLUS Software was showing fail status (205) with error code 206 (bit 128) - too hight load. We use Agilent pump - TPS compact…
Thanks for advising, do other errors show up in the logbook?
Please watch the movement of the sampler arm during initialisation:
- Is it shaky or jittery? If yes, turn off module and clean transport rods using IPA & lint free cloth.
- Which position…
Hello,
I will assume you mean you are using the backup restore tool for CDS 2.4 not the ECM XT import scheduler. If you have setup an import scheduler task to back the ECM-XT system up that would be self-referential and would not actually back anything…
Hello,
We have been having issues with our ICP MS performance and tune reports. Both have been failing with low counts. We have changed the peri-pump tubing, remade all standards, remade the tune solution, replaced the torch, and cleaned the cones…
Hi dubang69, Im also using export to CDF in OpenLAB 2.3 and OpenLAB 2.2 (with Data Export Addon). Unofrtunatelly second step for these revisions is to open CDF in some open source software as noreuter recommends. Im using lightweight utility UniChrom…
Hello community, I set up an Agilent 6890 (Software: ChemStation A08.03). A few seconds after I start the GC, the error message "Shutdown ( # 6 ): Back Inlet Flow Shutdown" is shown on the display. The settings of the Back Inlet shows that the pressure…
Hello, I have an issue when I try to install OpenLab CDS EZChrom Version A.04.07 The plan is to install EZChrom on a Citrix to allow the user to retrieve files from ECM for review or reintegration. I start the installation and select the Installation…
Hi all We are using GC6890N in our lab to analyze dry gas, the GC with TCD and FID and four columns Two weeks ago We noticed something strange in the results like a cycle. N2 value it goes up and then goes back down. Is there any advice to solve this…
This Information Applies To: Agilent 7683B Automatic Liquid Sampler
Issue:
O nly use this alignment procedure if the injector hardware configuration has changed or if the Align Mode light comes on.
Agilent does not recommend performing this procedure...
This Information Applies To: Agilent MassHunter data acquisition software and MassHunter Quantitative Data Analysis Software
Issue
This article explains the difference between quantifier and qualifier ions and provides some general guidance on quantifier...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Fragment Analyzer controller software
Issue:
Over time the dynamic coating polymer on the capillaries needs to be refreshed. Failure to do so can lead to a loss of performance. Indicators...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an...
This Information Applies To: Agilent 6200 Series TOF and 6500 Series Q-TOF Systems
Issue
Instructions on how to correctly dilute the tune mix for 6200 Series TOF and 6500 Series Q-TOF systems depending on source type.
Background
If the Tune...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0
Issue
The Agilent MassHunter Qualitative software can generate qualitative analysis reports, which may include sample chromatogram information and sample...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0 or higher
Issue
Identifying compounds using a spectral library is essential in qualitative analysis. This article will describe how to load a data file and...
This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
The Calibration Curve window tools help with setting an appropriate calibration curve for each compound in the calibration table.
The Calibration Curve window...
This Information Applies To: Agilent MassHunter Quantitative Analysis "Quant-My-Way" user interface (version B.09.00 or later)
Issue
To process the data files in the batch and obtain quantitative results, you must either create a new quantitative...