Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was?
From what I understand…
I have a 6490 QQQ. Each time I run either autotune or check tune, I get masses that are out of tolerance especially in MS2. The system seems to pass if the Dynamic Ramp Table enable button is checked. This is associated with the collision cell and detector…
Based on your latest results, I would say that nothing is fundamentally wrong with the optical unit in terms of noise, so I believe this is more an applicational problem - your passed quick noise test points to that direction. The error 1050 with the…
Disconnect the sample probe from the valve and start the leak and restriction test. If it passes this step, it means that there is some restriction in the sample probe, enough to cause that part of the test to fail. Try a brand new probe. If it fails…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
Seems then to indicate something in the flow path when including the mixer and damper. First, make sure that you do an elasticity calibration because you have changed important parts in the flow path. Then re-run the pump leak rate test with the mixer…
Hello experts!! I have 6460 LCMSMS system. It unfortunately failed in tuning even in brand new tuning solution. Auto tune repeatedly failed in peak width adjustment in MS1. What could be the problems and possible answers? Cleaning is done up to ion transfer…
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Fragment Analyzer controller software
Issue:
Over time the dynamic coating polymer on the capillaries needs to be refreshed. Failure to do so can lead to a loss of performance. Indicators...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an...
This Information Applies To: Agilent 6200 Series TOF and 6500 Series Q-TOF Systems
Issue
Instructions on how to correctly dilute the tune mix for 6200 Series TOF and 6500 Series Q-TOF systems depending on source type.
Background
If the Tune...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0
Issue
The Agilent MassHunter Qualitative software can generate qualitative analysis reports, which may include sample chromatogram information and sample...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0 or higher
Issue
Identifying compounds using a spectral library is essential in qualitative analysis. This article will describe how to load a data file and...
This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
The Calibration Curve window tools help with setting an appropriate calibration curve for each compound in the calibration table.
The Calibration Curve window...
This Information Applies To: Agilent MassHunter Quantitative Analysis "Quant-My-Way" user interface (version B.09.00 or later)
Issue
To process the data files in the batch and obtain quantitative results, you must either create a new quantitative...
This Information Applies To: Agilent MassHunter Qualitative and Quantitative Analysis
Issue
Calculation of signal-to-noise in MassHunter is done using the same routine for both Qualitative and Quantitative Analysis. However, there are differences...
This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software
Issue
Multiple samples can be run without user intervention using an Automated Liquid Sampler (ALS) in combination with a Sequence.
Users can perform simple edits...