At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
The Digest of upcoming webinars has been updated as of June 23.
Notable additions include:
Cell Analysis (1)
Spectroscopy (1)
Check Out:
Elemental Analysis in Yeast Cells and Microplastics by ICP-MS with Automated Micro-Flow Sample Introduction…
A 5% H2 blend should work also. If you're using multiple calibration standards, make sure the range of the standards matches or exceeds the expected range in your samples.
One thing to watch for: I have seen instances where H2 saturates the TCD at high…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and platinum shield plate must be installed after performing maintenance, or replacing with new components.
Items Required
Standard...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and shield plate must be removed and inspected during routine maintenance, and when troubleshooting plasma ignition problems. Faults with...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
This procedure tests for an air leak in the sample introduction system, or fault with spray chamber draining, by sealing the end of the torch with a piece...
This Information Applies To: 7700, 7800, 7850, 7900 Agilent ICP-MS instruments
Issue
Error messages related to gas flows can cause the plasma to shut off or disable plasma ignition.
The troubleshooting procedure in this article is applicable for...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Error messages related to water flow or water valve issues can cause the plasma to shut off or disable plasma ignition. While a warning message will still...
This Information Applies To: Agilent Fluorescence Detectors (FLD)
Issue
Sometimes it’s necessary to change FLD EX and EM wavelengths during a run since not all components will excite or emit at the same wavelengths. However, when the wavelength...
This Information Applies To: Agilent Analytical Scale LC Systems
Issue
To troubleshoot high backpressure within an HPLC system, it’s necessary to understand the expected pressure throughout an operating system. The total pressure for an LC is measured...
This Information Applies To: Agilent LC Samplers except dual needle multisamplers
Issue
High pressure in the LC sampler is sometimes due to a clog in the needle seat assembly, the needle, or the sample loop. The needle can pick up sample particulates...
This Information Applies To: All OpenLab CDS ChemStation Edition software-Standalone Workstation
Issue
After launching OpenLab ChemStation Online, the following error is shown: “ChemStation ChemMain has stopped working” . OpenLab ChemStation Online...
This Information Applies To: Agilent AriaMx, Software version 1.8 and later
Issue
The Aria Warm up Feature was introduced with Aria SW/FW 1.8 to reduce baseline drift. The warm-up feature is used to speed up the heating process that would normally...