Hi Evans! I would like to know if you ever figured out what could have caused the error with the oxygen sensor? We just had the same issue in our labs, and have got a new sensor. We are however hesitant to use the instrument again before we we have an…
Hello Columbia_METALab ,
I don't believe this was a known issue with Quant B.05, though it is out of support at this point. I can suggest a few things to try.
First would be general PC maintenance. Clean up Windows temp files and defragment the hard…
Hello drummar ,
I'm not finding this as a known issue.
For an error like this the first thing to try is the Quant Restore User Settings tool. Before doing this make certain you have any custom column or screen layouts saved in Quant. Also make note…
Hi Paul, I just got the chance to call Agilent. It turns out that our equipment and software might be too old to have this feature and we may need custom-build script to accomplish this. Currently we are using Chemstation. I will discuss this with our…
Hi yurisarmento
The system is behaving exactly as designed.
The line that has "Prevent analysis" stops the system from making more injections when a bottle reaches the value supplied. The "running out of solvent" option happens when the volume gets…
Hi Lstephan
The autosampler moves the sample backwards and forwards (draw and eject) inside the sample loop.
In the past, where I wanted better mixing than this, I would eject the sample into an empty low volume vial and draw and eject in that location…
Hi Daniel
The user manual for the Vialsamplers (I think that's what I'm looking at in your photo) has three options for this:
5068-0215 - Stator, Injection Valve, 600 bar/800 bar
or
5068-0216 - Stator, Injection Valve, 1300 bar
I'm guessing…
Hello everyone,
we have a LC/MSD iQ with OpenLab CDS. Recently it started aborting the sequence run. All the modules apart from the MS go offline for a short time and the run is interrupted. In the details I find the following error description: Aborted…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...