Hi Kenforce,
In this case, I would run the pressure decay test, as you could have a small leak that is causing the inconsistency. I would also consider increasing your oven equilibration time by a 0.5min or so, just to ensure the column phase is at…
Hi Marty,
We have a FID and an MS connected to the same system. We only want to acquire the signal of FID, but if we want to report the method parameters we are “forced” to acquire both (we do not really know why and we have already ask the question…
Sorry for the delayed response. I've found that either works fine. I typically try to keep injection volume low (0.2-1 uL for 1uM protein). I've mostly been using 50mM HEPES 150mM NaCl and get great deconvolution results, even with a 3 min short method…
I am getting message "draw command failed, EE34205" while running batch in G7129B infinity 2 autosampler. I have tried replacing transport assembly, but error is same. I have used another module G7129A Infinity autosampler with this LC, but same error…
Hi, I am getting same message. But auto tune and quick tune report both are very good. I am getting "RFPA difficulty" message during tune evaluation. How to solve this?
Hello all! We are currently familiarising ourselves with the MassHunter data aquisition software (LC/MS Data Aquisition for 6200 series TOF/6500 series Q-TOF. Version B.08.00 Build 8.00.8058.3 SP1). We have been trying to find the most efficient way of…
How do you set up an auditory sound alert for when a method run completes? I was thinking it may have been a runtime event but I I can’t see my to find it .I set a alarm on my phone for the length of the experiment to remind myself to go check and set…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...